For capturing pace maker pulses of 0.5ms duration, what is the recommended Data Rate settings for ADS1298 in LP mode? And in turn what will be the sampling frequency?
Regards,
Harish.
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hi,
For capturing pace maker pulses of 0.5ms duration, what is the recommended Data Rate settings for ADS1298 in LP mode? And in turn what will be the sampling frequency?
Regards,
Harish.
Hey Harish,
If you think of a pacemaker pulse as an ideal square pulse, it looks like a sinc function in the frequency domain. For a 0.5 ms duration pulse, the first lobe of spectral density falls under 2 kHz and the second comes in under 4 kHz. The ADS1298's fastest sampling frequency is 16 kSPS in LP mode. With the digital filter's magnitude response, this gives you a maximum bandwidth of about 4.2 kHz. It might be tough to get a good representation of the pacemaker pulse with this bandwidth limitation.
However, the ADS1298 has two configurable analog pacemaker output pins that will output a single ended version of one of the channel inputs conveniently for a discrete detection methodology. TI has two TI designs that cover pacemaker detection with the ADS1298:
Let me know if you have questions about these.
Regards,
Brian Pisani
Hi,
Since you mentioned sampling frequency as 16KSPS. please can you clarify difference between Sampling frequency and Data rate?
To my understanding if the data rate is 16KSPS we will be sampling at 256KSPS internally in LP mode. can you confirm this?
Regards,
Harish.