we are using the ADS1298 in an ecg-application.
DVCC: 3.3V
AVCC: +2,5V
AVSS: -2,5V
We have the situation that during test we lose some ic because of damage.
The effect after this happened is:
First thing we found was a peak on DVDD off 2ms to 4,2V during Power-Up. We put an 3.3V LDO in line to eliminate this peak.
A other thing that I can see with the oscilloscope is a short peak at AVSS to 0,8V ( it is coming out of the ADS1298).
So we put an Shottky-Diode in and we could limit it to 400mV.
Do you have an idea?
Hi Brian,
could you please take my problem serious. We lost two days for answering questins about informations that I gave you at the first request.
I did add a shottky diode. But that limits the peak only to 400mV. The limit in Datasheet is 200mV. The current, that is the source of that peak is coming out of the ADS1298.
What can be the reason for that current coming out of the ADS1298?
an an attachmentI send you oszilloscope-pic of the poweron with 3.3V(yellow) and -2,5V (green)
Thats the way we connectet the ADS1298 power-Pins
Thank you for your help
Jürgen
Hi Brian,
we have still the problem with damaged digital inputs at the ads1298.
If a input is damaged we can mease 500 Ohm to Gnd and 3,5kOhm to VDD.
Do you have an idea what could cause such a damage.
Jürgen