Part Number: ADS1298ECGFE-PDK
please also tell me how to remove dc offset which is about +/-300v.
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Part Number: ADS1298ECGFE-PDK
please also tell me how to remove dc offset which is about +/-300v.
Hi Ahmed,
DC offset voltage is expected due to large, imbalanced electrode impedance and also the input bias/lead-off detection current. IEC standards require that you can still measure the ECG signal amplitude in the presence of +/-300mV electrode offset voltage. You can either use the RLD to establish a mid-supply common-mode and reduce the differential signal gain, or you can AC-couple the electrode inputs.