I would like to measure the electrode impedance for all 8 channels using the AC Lead Off function. I have done some preliminary measurements and it looks feasible but there are accuracy and resolution concerns.. I would like to verify that the electrode impedance is below 5k ohm. Reducing the series "Patient Protection Resistor" would allow for more resolution/accuracy(?) in the measurements. I am aware of the patient risk current requirements. Is the need for this resistor related to a single fault failure in the ADS1298 that would result in AVDD or AVSS being connected to the patient input at the ADS1298? If so, are there ways to limit the current without such a high series resistance? Or, are there ways to limit the risk current without putting a high value resistor in series?