ADS1291: ADS1291 ECG signal is abnormal

Part Number: ADS1291


Hello!great TI engineer:

I have been debugging ADS1291 for a month now,The main problem now encountered is the failure to collect ECG signals from the human body.

I hope you can give me some suggestions and help me to locate where the problem is. I have almost no direction.

The product that is currently being developed is a wearable device that uses the ADS1291.The ECG signal on the ECG simulator can now be acquired.The ECG signal is as follows:The waveforms drawn using matlab are the raw data above and the filtered data below.

The following is the collection of ECG data on the wrist, no ECG signal,The waveform is as follows, and the original data is in the attachment. The waveforms drawn using matlab are the raw data above and the filtered data below.

875116
870047
864654
860525
858355
861531
863963
871701
877983
879098
875723
866874
859901
856767
855036
856359
862434
874834
878980
877596
871733
864755
855348
853579
852489
860268
867112
873997
876924
877099
869866
861514
856242
852023
851740
860124
866647
872963
876624
874838
870752
866335
859441
854402
852026
855938
863048
870552
876260
879316
875823
867873
863946
860143
855134
855943
866337
874650
878796
878166
871744
865434
859694
851541
851378
859577
868057
876310
882506
880804
872034
863762
858701
854459
856626
861534
867352
874723
877814
876764
874288
867904
862643
857876
857439
862549
867056
872138
880089
881201
877316
870809
862691
857968
855817
858037
868703
877661
882259
883904
878166
871541
864986
858131
858732
863069
871794
879142
881642
879924
874471
866093
860556
856249
854465
856866
862864
871095
876443
877714
874898
869434
864394
857336
855719
860900
868971
874667
881784
882196
878996
871636
864594
860014
857625
857943
867443
876706
880943
883898
879353
872743
866778
860131
857569
861084
870030
878264
881489
881945
878020
868124
860267
859214
860521
863326
869893
876831
881462
881652
877019
868991
862363
858440
859532
863828
870115
875256
880610
883427
880814
873781
865835
860698
859458
862804
868704
876732
882873
885119
882813
877013
868747
861493
858590
864746
873394
882761
889182
886248
881013
873859
864351
862604
865999
872298
876180
879718
885183
887023
881795
875093
870202
867786
864786
868478
874747
881311
884318
885606
883679
877723
870111
866181
864134
867316
872504
880353
886264
890301
888595
880272
870449
864746
860459
865664
873214
881721
888634
891726
887569
879140
870884
865844
865289
869088
875203
883306
888634
890339
887715
877778
871171
869901
867331
868813
876713
885845
887751
889652
885575
878778
873464
866973
864037
866506
872959
879036
886091
891441
887814
882104
877413
870943
867213
869676
876397
883199
890104
892444
888429
880608
872877
866907
865464
869599
877352
883419
887625
890373
886199
878157
873396
868275
866452
869411
874660
880956
886078
886724
886561
884013
877293
870240
867833
869291
874026
880262
888764
891954
890178
882403
875708
871066
869433
870658
877101
884223
894146
896571
891178
884503
877208
870386
870126
874423
882731
889379
894798
896908
892268
881356
876714
871258
870610
875102
884268
890737
893426
893895
890421
884004
879059
873754
870808
873878
879813
885655
893014
896447
895839
888524
879944
874761
875182
875334
880886
890127
896949
899926
897767
891058
883850
876061
872524
876201
884836
893231
900243
903046
901587
894579
886944
883329
881171
884580
890901
896713
901828
903169
898124
893551
888563
882899
879171
880853
885629
894619
900103
902685
902263
897368
890229
883093
879811
881424
886629
893424
899743
901810
899829
895181
889521
884085
879611
880578
887196
898499
906013
908504
906004
897650
887449
882354
882009
888369
894346
901848
907461
910985
905826
902608
897682
890208
884567
886240
893710
902812
905063
907775
908607
904298
897665
890932
883800
880613
886086
894815
902092
906791
906703
901218
894886
889715
884824
886064
893951
904604
911417
913867
910591
904375
896826
890786
886840
889085
898023
906061
911681
912984
910176
903196
896284
891075
889596
890908
895159
902915
907206
909630
911675
908913
903394
895741
888543
890201
895618
906511
918389
921683
917814
910554
903065
898350
893726
891740
898227
906854
914171
916002
916370
911108
901961
892960
890842
896203
903846
911557
915719
917968
915181
907670
901103
895326
893427
893928
900277
907738
913916
916539
916250
912137
904759
896421
892577
893251
899933
907101
912537
916806
918169
911176
904303
896067
890245
889160
895210
904053
914032
918144
915911
909591
903958
895863
893490
899041
906304
911425
917339
920372
918358
911460
904138
898463
896614
895436
898614
907144
915535
917250
916916
914038
909818
901681
895560
898283
904804
909609
914249
920611
926728
926346
919410
912233
907964
906424
910988
919454
927043
931133
928591
919850
911474
903561
897293
896951
903101
913164
917986
919961
916150
911984
903693
897861
894021
894886
901861
909876
914308
917363
915596
910951
905076
899654
894416
892922
897428
901926
907714
913398
915336
913663
908015
902810
897369
894466
897206
906099
911271
915697
915755
910687
903044
897205
892339
892491
897251
905233
913526
916553
915656
910585
900286
893139
889456
892085
898594
902271
905761
913833
913000
908019
901758
895160
889611
889753
895341
903377
907413
910799
913060
910708
903798
896239
890527
890278
893475
899717
908667
914658
915698
912396
907034
900005
892583
889077
894073
902455
911549
916190
915900
910800
901946
892880
888074
886924
894658
903566
911391
915475
914493
906623
899589
895268
892438
892638
898106
903447
909409
911072
913523
909663
901288
895819
891931
890771
894380
899715
908364
914528
915639
910928
904045
895677
891286
888651
893178
901344
909608
914940
914742
910224
902993
894062
888938
889087
895289
902782
910116
914554
916704
910084
904347
897389
892499
892933
897178
902534
907879
915182
912612
908054
903822
897542
893686
893196
895227
901354
909595
913525
915605
909440
902091
894902
890644
889406
893207
901011
909030
913849
914662
913520
905364
895901
889182
890561
896615
904167
909827
912947
914331
909620
901099
895436
891973
892000
897161
905352
910976
916850
916299
910156
906240
900054
893852
891318
895371
901689
907709
914646
917269
913576
905117
897044
891118
889691
891394
899646
909828
916606
916911
912858
905746
897960
890108
890141
894958
902317
910825
915236
917088
912154
901726
896931
893375
890956
895854
903556
909629
913899
913453
910424
904333
897416
892218
892624
895184
901534
905364
911323
915129
912639
906084
897773
893922
893346
893506
899586
906423
913744
917081
914182
904526
896034
890403
888400
890986
898623
908715
913276
913909
914216
904342
894492
889185
889506
893101
899618
907797
914828
914605
908717
903746
898144
893695
891354
894369
902872
908848
913503
915464
912039
906288
899594
893330
890698
891716
895809
904648
915014
920919
916636
908436
902239
894121
889373
893535
900081
906896
913058
914636
913544
905279
894295
890011
890541
893919
900661
909149
913793
913624
911370
904336
897073
892681
891598
893661
897831
904753
909415
912194
910204
906564
900306
894508
892394
894965
899113
907796
913977
918172
916219
908194
901293
895558
891807
894976
900051
906918
915260
915946
910592
904201
895168
890305
890046
895638
901564
908118
913910
917227
913660
907297
899698
895501
892943
895748
902235
909974
912229
910836
911490
907598
900494
895738
890981
891388
895524
903146
913280
917179
915998
910586
902060
894849
890176
891143
899306
907369
914654
918556
915852
909614
901068
893741
893375
899159
906142
909488
913459
917951
912835
905428
897771
892582
893017
893868
899791
906760
912498
913087
912068
908274
902538
894711
892078
897364
901661
904303
909529
914465
916068
912329
903608
895375
891573
890851
898574
908400
915758
919622
918104
912894
903791
894822
891771
895093
901388
909806
914591
916673
914076
907138
900642
895560
896828
899901
904321
909208
915629
916833
915819
909853
904308
901158
897270
895889
901548
906674
912341
918069
918336
911931
905966
901576
897176
895611
902007
909441
915153
918669
917444
912266
906761
899174
893619
894756
902843
912268
918245
919486
917731
911314
902641
896781
895836
899959
906654
912154
916781
918790
916724
913779
907573
902265
900152
898283
903241
909978
917638
921243
920581
915981
908024
902272
898819
897359
901233
910306
917716
921978
918373
913189
909276
903436
898046
900045
904945
912411
918713
923548
921501
914697
908371
901256
898830
901127
905364
912130
919261
921872
920623
915899
907734
901411
898783
901218
906463
912143
914778
919287
920137
916671
909521
902535
898925
899052
901858
910851
920194
925359
923412
916497
910586
904844
899015
898395
904421
916036
922046
924899
923803
918319
908538
901333
898177
902261
909984
917121
922577
925089
922126
915561
909048
905207
903998
902618
907384
916646
921111
924450
924875
921031
915004
906843
902744
902811
906019
915139
924564
928805
927594
921182
917876
911989
904173
902213
910419
918203
924115
928916
927278
921750
911640
902566
902947
905810
909161
917858
925968
929546
929168
921589
915004
907856
907384
906698
911563
918879
923289
925451
928138
924544
918916
911584
906710
907288
912958
917824
925921
930431
929609
924363
917931
910546
905888
905578
910833
919260
925527
928518
929789
926018
918503
910439
904553
906446
914546
920167
926148
931073
930074
921857
914544
908348
905870
908968
914126
918570
923374
925044
924935
922171
916991
909602
905262
905546
911609
915623
921041
927033
927908
924298
918318
911308
907583
906160
908923
917892
926001
930566
929273
924314
917606
911195
905811
904938
910556
918101
924148
927233
926940
920733
911920
909126
907588
908056
911009
915708
922638
926707
926081
922547
916264
912406
906124
904473
910912
916514
920591
923694
925330
923029
915411
911591
907360
906916
907527
914803
923596
930130
931001
927296
918978
911038
906077
905428
907708
917421
926052
928690
927315
923267
915126
908657
904442
906249
911041
916733
924258
928480
929306
922886
917502
910794
906664
908495
909481
914940
919616
924518
926234
921278
917076
910868
904480
903277
906281
913272
922193
928758
929166
925574
918258
910534
904883
905052
908909
915948
921611
925694
928013
925755
919053
908713
903459
902876
907398
914374
920198
925738
928073
922271
915504
907986
904561
902166
905427
913026
919547
922764
924903
923281
917601
911751
905721
905824
910803
914721
922218
928766
932599
928171
918131
909226
903239
900256
904616
912688
922209
927839
929584
924302
917831
910021
903595
902075
908464
914531
918686
921675
923284
917138
910448
905517
901247
899807
904446
910969
917999
921632
924418
922056
914288
909380
903963
902264
905529
909188
915586
921996
925709
921548
912942
906781
902801
900275
903181
911634
917783
922851
924316
919476
914354
907462
898918
898834
904094
911638
917783
922214
924072
921117
914241
906894
902003
900224
903455
910138
918292
922710
922130
921046
915889
909142
903857
902444
906171
910943
916246
923941
927875
924959
917562
908791
904766
902562
904361
912438
918308
925118
926999
923273
916711
910233
903024
902939
905604
913785
921918
926186
926836
923698
915487
910966
906853
904611
906584
913249
919661
922814
923453
921710
918696
913264
905681
902885
906649
913341
917287
923368
927149
927438
924036
914897
907046
905994
909511
915079
921506
929516
933358
931141
925146
918102
910023
908003
912703
921969
930749
933131
933646
931971
926611
920346
913753
913706
916179
920964
927553
933189
934933
932879
928343
923763
916686
913171
915774
920956
925920
930576
936352
935483
928962
921761
916033
912214
911591
916463
928204
935389
938631
936067
928874
921596
914809
910878
915245
922052
929715
934142
934990
932967
926444
916542
910010
907818
911953
917449
922339
927403
926980
922242
917777
912684
908351
901708
899034
903677
910802
912550
915224
913575
908373
902375
895473
891187
892623
896996
901139
906865
909421
906292
900103
892946
886811
881610
881406
887635
896833
902578
905927
905860
899413
888700
882818
879874
882964
889021
894448
902158
905686
902129
895339
887793
884133
882751
884198
890207
896333
899980
902751
901365
896571
890918
883678
880424
880869
885350
890986
897692
901082
900514
896152
887416
879631
876346
876353
882896
892528
900019
902561
901769
896205
888783
880164
878773
880905
885191
891609
897503
902080
900896
894866
889921
884414
882588
883556
887141
893961
897924
898058
899677
896748
892667
887130
883549
884672
888549
892684
899560
904176
905529
901299
895782
889808
885718
883608
888869
896109
902564
907784
906403
903324
896182
887486
884358
886822
893468
899914
903156
906223
908076
901707
895751
890234
887168
888463
893061
899796
905284
907308
905698
904314
900489
894983
891322
892430
896395
902237
906744
910493
912858
909350
900735
892102
891307
892217
895962
903499
913577
919268
918376
912924
905114
895236
888595
889461
896936
906399
911187
914626
915321
910565
902656
899066
897403
897670
902971
910009
913947
917461
915433
911758
906849
900726
898249
898229
904068
909479
914889
919753
920684
917556
909713
901013
897183
897213
900934
908848
917761
924019
924479
918763
910821
904570
899008
900193
906854
915961
923541
925186
923803
919423
909050
903447
899706
900897
908730
917124
922108
925720
923451
918158
912522
908622
905313
905326
911061
916954
921328
927375
929001
925750
918936
910399
905710
906426
910366
916896
925770
930939
931801
926479
920439
913293
906748
908054
914343
923986
931201
935566
936006
930959
923071
916118
912213
915508
921494
928495
934569
936950
934936
929411
925211
920138
916946
916403
922671
928765
933936
936844
938908
935626
930672
923976
918912
919354
922459
926021
935317
941652
943011
941050
933118
923201
917719
915144
921247
928914
939633
944929
945006
940283
932383
922509
920068
921831
927581
934123
939963
946331
946491
937067
931414
926751
924293
925199
929157
935522
943984
947733
948338
943408
936108
930623
926414
923897
925746
930556
936664
941954
945009
941300
934895
927838
923333
920744
927088
936123
944564
949621
949209
943666
938021
929294
923831
925373
929223
936511
943190
947939
949499
943305
935351
929859
926144
926857
932108
939143
943848
947839
947441
942766
937826
933453
928898
927116
930858
936533
942602
950261
951698
948069
939488
931748
926781
925626
927204
934452
941043
948602
950813
946913
940561
932598
925373
924161
930389
938124
944094
947552
948173
946035
939049
933077
927228
926384
931178
938164
944001
949084
949724
946570
940997
935186
929597
930640
933052
938366
943982
949986
952721
949968
943014
935582
930040
927669
929424
935739
942691
951184
954087
950014
943084
934700
927368
928464
932621
939366
948213
953513
954828
950459
941194
934273
932336
932901
935144
940099
946794
953926
953858
950508
944186
939418
936992
934774
935006
942969
949873
953835
956149
954378
949085
941954
936054
934650
937999
940692
948046
955760
958384
954037
947401
940511
934253
931257
936654
945833
954383
958969
961723
956419
947948
939132
935545
937036
941986
946465
954122
959274
958407
953114
948135
942723
939334
937713
939641
945761
952018
956090
957673
953346
950356
944131
938146
938201
940266
941788
949563
956349
960719
959085
952379
944837
938249
933764
936704
944818
953754
958956
958740
956863
954245
943669
937296
937306
942754
949332
953956
959393
962724
957747
950696
943239
938821
938206
942068
948374
952459
956505
959358
958193
953475
946399
940002
938761
941945
946716
952852
959277
963842
960686
952695
945438
941043
937216
939003
946911
954552
961434
963814
961553
953686
946009
939168
938377
943078
950064
957234
961828
964405
961115
952953
949238
943430
941330
942926
947538
952481
958276
961083
958924
955648
950688
943550
940229
942237
947694
951652
958751
963395
963159
957727
951806
944818
941468
940762
946162
955849
963190
965158
962313
957072
949424
940806
939648
944079
953171
960476
966168
967559
964150
956173
949934
946501
945308
946312
951746
958835
964218
965029
961382
956243
952980
949259
945624
947260
953318
959044
964478
967448
966217
961512
954806
948978
945319
946026
952402
960376
968769
971497
969426
962743
955016
947638
944120
948017
957133
963958
970377
972736
970553
964274
956321
951111
949395
952001
957151
964724
973813
974251
969147
965104
960237
955138
953389
954762
960640
968779
970865
974139
972457
968234
962001
953943
950878
953958
957559
965508
972771
976438
975681
970518
964222
956991
952075
953163
959158
967783
975584
979796
977121
971551
962172
958058
956093
958983
964735
970696
975665
979171
976893
972713
966076
961187
957533
958389
965726
972167
978419
982544
982490
978370
968791
962917
960055
960421
963881
971803
978922
983634
982181
975431
968208
962736
956288
957144
964872
972554
979170
981514
981338
977839
969114
962463
960359
963676
968666
974873
979378
982476
981043
976321
970904
965421
961699
962340
966896
975623
981863
982741
980948
976908
972746
966468
961406
961589
966955
973589
980476
985364
984076
976346
969699
965855
962954
961890
967798
974530
980691
985412
984486
978771
972487
964774
961764
962384
966148
972946
981253
985103
983195
975458
970824
966384
963330
962676
967763
975238
979969
981946
983227
980031
974409
968220
963511
963650
968767
974149
980843
986209
987963
982893
975071
968133
965071
963814
968283
974868
983516
988754
988132
983351
975704
968598
964605
967261
974478
980299
983305
987951
987051
977517
968306
963018
961811
963139
969161
979204
984586
986276
985544
982564
977730
973675
971165
970759
973198
979374
984893
990006
991518
986297
978236
971406
967464
965781
969788
981007
987449
991548
990512
986772
979973
972951
968227
968226
972367
978920
985262
991816
991578
986859
977747
970634
967009
969380
975228
982189
987113
989829
989039
985776
979672
974259
969842
969046
972812
978229
982589
987132
989590
989352
982144
974896
969484
970122
972553
978529
985293
991527
993453
987770
982031
976483
969634
968317
974253
983501
989456
993629
993123
987798
978724
971851
970327
973819
976774
981504
988872
992396
989921
983069
978339
974942
971301
970818
973646
981496
986781
988881
989321
986112
982331
976711
970780
969181
971969
977873
987123
994430
996212
990852
982576
975340
970014
967255
972129
980518
988188
993736
994138
989011
981693
973744
973506
973782
978673
983534
989311
996730
996499
989089
983024
975696
971832
971783
975946
982136
987197
989600
991569
990713
984909
977616
971706
970711
975735
980067
986396
992543
996129
991406
983489
976368
970977
967873
971761
979168
987205
992489
994230
991448
985849
975359
969944
969252
973531
979097
986073
993064
995333
988326
981444
976068
972299
971073
976569
983907
989034
992229
993052
990078
983781
977300
972280
970476
974789
979169
985407
991228
994238
991519
983610
976615
971996
969148
971499
979479
988474
993021
993562
990899
985820
978348
971633
969671
973928
981516
988661
993358
993461
988798
982495
976181
973057
973755
978168
983280
989593
994136
994564
991765
987325
981609
974460
972399
977148
982876
985883
992396
996641
994779
987997
979669
973859
972513
974219
982622
990289
996020
997069
992199
984971
978983
973038
972200
977002
985004
992092
995693
996886
995884
988006
980242
976450
976321
979909
985051
990265
995349
996731
993914
988564
983966
978944
978117
979749
984672
988496
993741
998159
998399
991392
983587
978588
977481
977938
984059
992853
998544
1001411
1000251
994439
986241
977842
974123
977771
985671
993804
1000421
1001253
999993
992153
981979
978334
978284
981786
986994
994002
1000039
1001846
997903
993541
988025
982734
980555
983361
989279
994616
998308
1001228
999837
994880
988220
982598
980521
982911
985651
994906
1002555
1004518
1001711
994593
988296
982816
978772
983711
991836
997008
1000598
1002701
1001726
996790
987793
981681
980359
985139
990180
996154
1000729
1003790
999318
993709
988259
986021
984184
985605
990466
996617
1000798
1003984
1003202
999028
993143
987104
983616
983640
988551
996866
1003576
1008454
1005856
998003
991980
988094
983545
983289
989864
997779
1005133
1008297
1004518
999309
991954
983283
981871
986789
992188
998241
1002817
1006855
1005136
998510
992829
988677
986522
987175
991805
998408
1002333
1004439
1005853
1001162
994968
987433
983893
987537
992294
996461
1003378
1008781
1008420
1001954
995604
989884
985353
985053
992341
1001784
1009533
1011695
1009402
1003557
993671
984895
984287
987661
997900
1004832
1007571
1008319
1007016
1000779
994762
990754
989389
989701
996443
1002177
1006899
1008305
1007603
1002491
997684
993196
989322
990656
995600
1000801
1007293
1010413
1009943
1004893
997138
989674
986917
986727
993223
1001033
1008297
1012408
1013308
1008001
999198
989973
987669
990134
998625
1005323
1011099
1014639
1011098
1002548
995465
991258
991572
993743
997464
1001261
1009109
1013011
1010291
1005835
1000456
995356
991362
991808
998093
1004154
1009835
1013676
1013151
1010067
1001743
991798
988812
987184
991075
1000522
1009353
1013585
1013651
1008382
1000273
993806
988393
989861
997889
1006168
1011302
1015497
1015614
1010075
1000296
993646
991953
996038
1001076
1005160
1010450
1014233
1011836
1008637
1003176
999188
995189
995115
998589
1003423
1008143
1012711
1013474
1009809
1001886
997230
992689
992743
997601
1004353
1012251
1018089
1015486
1010424
1004318
996242
989958
992091
998648
1007379
1012711
1016226
1015428
1010334
1002269
996718
993294
993912
1000586
1007384
1013453
1017168
1015580
1008657
1002239
998229
996592
997607
1001719
1006943
1010993
1013462
1014071
1009578
1002597
997734
994100
994694
998926
1004143
1011409
1016189
1016607
1012694
1006853
999928
994861
993466
999007
1006496
1012961
1017119
1016193
1011388
1003674
994878
992827
996286
1003823
1009384
1013988
1018698
1017267
1011181
1006039
1001048
999025
998224
1001268
1007092
1012384
1015943
1016620
1014338
1009276
1001419
996431
996989
1002274
1005441
1013476
1018081
1018584
1013501
1006736
999008
994023
993340
999511
1007933
1016490
1018495
1016957
1012396
1006969
999796
994929
995536
1002773
1008916
1015061
1017294
1015270
1009576
1003768
1000272
1000052
998211
998825
1006333
1012819
1016059
1013863
1010023
1006841
1002522
996173
993758
997882
1001929
1007575
1013476
1016140
1012681
1003673
996184
993908
991152
994309
1002454
1009503
1014147
1015583
1011407
1003929
995074
989996
990549
997073
1005000
1011603
1014140
1014494
1009496
1001860
996217
991935
992546
997239
1003014
1011308
1016299
1012867
1010282
1003958
997196
991534
990793
997311
1003029
1006852
1011989
1011859
1010324
1004263
997908
993709
991728
995383
1000544
1005716
1011311
1013344
1009861
1002871
996174
991622
990438
994127
1001119
1007299
1011920
1012254
1008087
1002094
994971
990554
990205
994532
998247
1005638
1011363
1010788
1007963
1002273
994364
991482
992348
995273
999444
1007132
1010385
1009699
1007071
1003638
996622
991733
989434
991549
998268
1005285
1010115
1012409
1009072
1001189
993023
988596
987961
991981
997834
1005294
1010492
1012613
1008633
1000995
992324
988428
988403
992904
999535
1006419
1010218
1010878
1007390
1001151
995503
991850
991101
995258
999659
1003539
1008095
1011928
1008858
1002906
995129
989884
988173
991787
998646
1004316
1009838
1010516
1006727
1002111
995288
989096
987998
990879
997818
1005726
1009572
1009733
1007939
1002801
995705
991098
988728
991835
999623
1006206
1009381
1010699
1007316
1002246
995834
992312
990064
991721
997649
1003244
1008704
1011114
1008325
1000833
994074
989969
989230
990978
996858
1001446
1006525
1009781
1007716
1000823
993146
988608
987726
989608
997053
1004557
1009714
1010442
1008064
1000524
993014
989429
987866
990963
999030
1004109
1007806
1009993
1008284
1002179
994694
988694
986251
989111
996674
1003920
1010354
1012129
1007984
1000178
993812
987443
984393
989509
995481
1002763
1008597
1010452
1006825
999921
994228
988451
985133
986921
993501
1001369
1006653
1008066
1006368
1000425
993086
987438
984762
986176
991336
998256
1003789
1007135
1005155
1000168
993581
985946
984459
986708
993184
998567
1001336
1004662
1004330
999288
992318
986791
984594
986353
991516
998724
1003788
1005609
1002698
994606
988012
984083
983613
985153
989564
997324
1002378
1003579
1002838
995282
988015
983114
979539
981763
988599
995558
1000563
1003809
1002209
997010
989505
982043
979862
984174
990759
994696
999683
1002593
1000109
995037
989976
984218
981786
985201
990728
996596
1000911
1002980
1001168
996870
991573
986084
980789
984749
991004
995604
995919
996679
995654
990403
984264
980118
978249
979842
985886
993318
998054
1000800
1000202
994447
988276
984184
983034
983398
987114
992353
999164
1001377
998924
993391
987734
981867
979546
982829
989061
995006
998727
1000216
999714
994863
987416
981131
978047
980088
985332
990299
997016
1002451
1001006
993409
984983
981459
978247
977415
984291
992411
998496
1000563
998268
993674
987174
980836
975528
979474
986915
992735
997106
998446
995315
991234
986848
978807
974920
979571
986938
993016
996791
998994
998924
994752
987835
981114
977829
979468
985929
991779
997036
999536
997911
992150
986529
979783
977169
979686
984287
989359
994208
997266
997436
992946
987124
980306
976266
978484
983856
990624
996516
998344
996282
993728
988833
979789
977661
979193
982453
989218
995669
999344
997998
993185
987276
981770
979012
981842
986164
990714
995499
998222
998271
994116
986935
980263
977789
980427
985144
991036
996536
998470
996059
992322
987840
980905
977781
978716
984086
990385
995359
997842
996353
992154
983689
978006
975091
976711
982085
988566
992289
997239
998033
990841
984157
979535
974980
975522
979924
987887
992816
997179
996558
991060
983914
977340
975066
974616
978369
986034
993794
996473
995623
989663
982591
975046
971269
973451
977596
983685
988946
992573
991639
987772
981362
976117
971296
971878
977003
984233
989809
992201
991051
987104
980660
974308
970211
970309
975607
983523
989694
991913
990283
986592
980847
974320
971860
973269
976921
983805
990579
992630
992235
987525
980441
973312
969774
970244
975985
981094
987579
991706
990500
985420
979311
974453
970150
968305
973715
980973
984994
988135
990016
988059
981263
973812
969918
968824
973609
979302
986038
989468
989086
983548
978208
973164
968734
970499
974502
980101
986926
990643
988709
983177
978078
972460
967676
966982
972223
979994
987140
988306
986971
984843
981461
974588
967898
968028
972924
978066
984456
988049
986809
982918
978194
973594
968473
966300
970778
977394
983134
988295
987933
981450
976087
971548
967563
966588
969337
976170
985254
989772
989303
984939
977544
971081
966503
966103
971314
979113
984793
987698
989359
985201
978491
970090
965478
964619
969480
976216
981740
987073
988093
982821
976156
968908
965647
965016
969609
975720
983276
986749
986178
981471
974317
968857
964548
963322
968237
975681
984183
987101
985338
981690
976608
972218
967833
965532
968369
975834
983350
986273
983773
980811
975616
970849
967780
966080
968569
974374
981116
985423
986871
982915
975863
967140
963486
965316
967749
974194
981096
984468
985433
982314
975820
969213
965287
963828
967901
974532
979987
984536
984406
979961
975474
969401
965042
965338
966676
973267
979691
986007
986326
982174
976108
969483
963931
962015
965878
973086
980026
981222
979201
978165
973991
970858
966828
965353
967481
972072
978421
984266
984478
980332
975593
968764
963523
963408
967889
972736
976743
980331
983192
979979
974040
968506
964173
960464
963991
972255
980028
986328
987327
982189
975672
968819
964373
963536
966454
971203
977030
980446
983681
979784
972454
966813
964046
963251
966882
973541
979366
981934
982564
980129
975476
969426
963230
960666
964055
971517
977327
982189
983266
979593
973572
967449
962038
960388
964291
971313
979761
983748
985736
982944
976558
968698
963775
962021
965513
972497
977631
980749
982190
979808
973209
964876
959633
959749
963880
969899
975514
980738
982918
979763
976414
972046
964919
962516
964954
971105
976324
981476
985370
983354
975864
969162
963785
962234
964626
969026
973906
978833
981220
980655
974944
968029
963161
960696
963929
970281
975731
980739
984229
980545
974657
968948
964008
961533
963860
968994
975079
979621
982080
980403
975414
968926
964403
961806
963177
967019
973133
978742
981323
980336
975931
968414
961884
960346
964091
970366
975760
980188
980397
980213
974381
967124
961527
960449
964416
970494
976249
980759
983129
981283
975217
968169
963675
962250
965624
971529
977309
982610
985258
981767
974904
968871
964736
960985
963539
970194
976466
980861
982539
982081
977006
969576
965339
962778
963678
969292
974848
979774
983555
983642
977955
970899
964834
962458
963490
969443
976566
983171
985006
983426
978197
970795
966044
963704
965779
972247
978519
983084
985898
986096
979624
972412
966752
964550
965174
970921
978571
985185
987466
983516
976661
969776
966110
963663
966868
973534
977951
982876
985616
983278
977905
972185
966449
963594
964884
969974
975036
981284
984451
981763
977449
971756
965691
963646
965297
968467
976029
984373
987084
984464
976357
970714
966124
963468
965131
971013
977721
983220
987029
983218
975774
971882
968909
964955
965436
970294
975294
981941
984768
983313
978108
972693
966253
964419
965868
971261
976211
981124
984931
984529
978256
970693
966067
964508
966026
971237
977095
983469
987184
985981
981363
973144
966509
965313
965910
971404
978869
983476
986030
984303
977502
971759
967316
964581
966876
970405
975686
981918
984915
984429
980224
974089
968533
964423
966447
972188
977004
980881
983327
983396
980619
974785
970548
967708
965951
968593
975198
982481
987099
985717
980507
975863
969922
965151
966381
971003
977714
983676
985759
985196
979744
973079
966103
963884
965216
970149
977743
984246
986204
984816
979323
973630
967571
963887
966160
974240
979430
983558
986487
985911
980709
973439
967596
965506
967366
973233
980377
986548
987594
985558
980229
974526
969983
967406
967847
972727
978114
983733
986671
985348
980074
972825
969039
966519
967927
972431
978164
981559
983536
984579
980601
974258
968560
965664
967674
973066
978089
984803
987921
987154
981779
975031
968745
965047
966530
972203
979367
984772
986214
987271
983361
976081
970273
966508
968823
973748
979041
983727
987460
985023
981168
976158
969001
967136
968235
970324
978411
985687
986139
985407
981879
975704
968505
964276
965360
968052
974938
982822
987079
987486
982756
974767
967043
963429
965025
969416
975891
983978
988211
986894
981025
974229
969560
966108
966505
968659
973178
978154
982227
983986
980343
974844
968918
964662
963032
967013
974702
981801
984882
985474
981194
974266
967513
962769
962008
966138
973006
978485
982606
983808
979810
973953
967127
961959
962543
967446
971980
977971
983365
985411
981053
973404
966029
961714
962954
967564
973383
979983
983954
983755
979718
974648
967524
963188
964889
968583
971963
976675
982628
983101
980009
973709
967943
964546
963511
966421
971177
977372
982455
982277
977670
969631
963954
961879
963493
966878
969914
976083
980943
981585
978220
972743
966494
961880
962431
965831
971104
976618
981476
982822
979580
972727
966444
961874
962170
966370
976068
981446
983823
983161
978848
970521
963129
957931
958306
963936
970848
977057
980584
980398
978131
973719
967597
963974
965504
967881
972682
978136
983186
983894
979335
972786
966492
963339
962236
965426
971049
977213
981983
983096
979148
972123
966886
961201
959228
964707
967341
974989
981870
982653
978591
973818
967724
963113
961118
965152
970471
976051
981819
983286
979512
973359
967752
962167
960061
961993
967959
975106
978490
978874
975634
967568
959128
956188
956475
959123
963461
968993
973143
975819
971038
965223
958803
954252
952044
955892
961094
967096
972247
973930
971525
963731
957112
952144
951648
956680
962195
967846
972593
973708
969426
963387
958520
954022
952945
955196
960735
966262
971220
973658
968389
959289
953356
951749
951788
956301
962801
968291
971356
972429
967568
962199
955683
950078
949431
952639
960146
965952
969605
972883
970279
964563
956346
950748
951174
955952
961621
965572
969849
971058
968065
963886
956856
951593
951060
952831
959329
964614
968826
971014
967495
960382
953516
949059
949424
953795
959375
962524
966555
969931
967196
961286
954008
949158
946230
951113
955796
962043
968429
970574
967943
961466
953836
947511
946040
949535
955132
959104
963829
967168
967900
961595
954062
948699
947066
949755
955733
961656
966555
968678
965582
958356
953164
949734
948165
950288
955481
961423
966855
970668
967422
959891
953238
948886
947506
950936
956100
962498
967436
969459
966033
960141
955708
949599
947689
951296
958023
963013
968235
968444
966040
962048
955403
950929
947798
950586
956638
964332
968550
968009
966386
962565
958170
952873
950798
951925
956836
962534
967699
970685
967564
962389
954201
949643
949991
954428
960965
965980
968109
968173
966518
961719
955773
950491
948753
952536
958041
962491
966865
969329
968118
964841
958091
951972
948218
949296
955889
962446
966466
968959
968064
962376
956773
952771
950333
951907
956604
960925
966218
969400
967737
963288
956598
950347
948706
951202
957135
962916
967521
969899
969393
962892
954570
948913
945826
947492
955446
961586
965764
967997
965739
961638
954250
948294
946134
948380
954428
959749
963058
967234
967466
962404
954363
947239
944727
948196
952113
958913
966319
970511
968383
961991
954766
948927
947159
950013
956773
961288
967021
969306
967358
960991
957069
952667
948568
949488
954259
959418
964677
967992
967248
964359
957369
949987
946741
949156
956258
962746
967148
970171
968984
962569
954633
950459
947287
948782
953849
960401
965796
969716
968307
963034
957494
951479
948913
950634
956294
962703
967740
969918
968438
964676
957728
948259
946110
949486
958526
965850
970895
973759
972739
965079
958121
954026
951638
952114
957557
962951
968103
970271
968723
964133
956741
951531
949239
950064
956049
963626
969184
971417
969246
965098
960101
952708
947993
950493
956159
962698
967908
969501
969901
966778
960442
954666
952537
953829
958543
963852
969812
973486
972898
967216
961686
955418
953182
952380
957023
963997
970581
972090
970887
967839
963653
958157
953788
956315
960068
966377
969695
971935
970323
967606
962578
956856
953313
953974
957613
963451
970449
974627
974709
969235
963879
957172
952976
955286
959788
965552
969825
974374
974178
968409
963014
958304
956064
957446
960966
965984
972310
976066
976747
971913
965809
958852
956221
957801
964139
972134
976931
977339
976274
971783
963051
956721
951863
952796
956862
963743
970199
974858
975233
972119
964387
957667
955754
955953
960208
967554
974991
978909
979046
974065
967091
960199
956134
958100
963926
970920
977326
979907
978389
974883
968564
961992
957084
958508
962092
967660
973859
978098
978889
974080
966952
960678
957569
957761
962909
969762
976237
978751
977867
974763
968371
960953
955634
954324
960999
969003
973756
977856
977898
975024
971747
963009
957958
960376
964382
969498
974479
978461
977194
972939
969330
964832
960680
960606
965164
969838
974916
982249
982461
976963
970607
964790
961838
962283
966071
970071
974180
976498
978979
975819
972221
967699
962097
960326
965302
971273
976770
981116
981721
977255
970648
965271
962522
962014
965581
971516
978451
981799
983661
979934
972143
964904
961824
962862
966063
970213
976456
982624
984429
979666
972214
967155
963329
964221
966861
972173
977867
982779
985721
982778
976381
970291
966232
965103
969322
975517
981401
984804
985381
982116
977306
971111
966738
967359
971851
976109
981619
985979
987384
982787
974543
969456
965416
963385
967178
975722
981908
986104
986821
982803
975938
970805
966472
965587
969767
975268
979971
984113
984476
981058
975211
970293
966049
965646
972038
977974
982329
984738
984361
980301
974323
969543
966026
966201
969793
974749
980938
984108
984273
982113
977323
971161
966233
963174
966537
975124
982068
984949
985404
984000
980498
973979
967798
966406
971163
976966
982263
984638
985415
983083
977283
971908
968557
966796
970863
976345
982173
984021
981428
979866
978701
974173
969353
966414
969464
975849
980078
983763
985542
983548
978826
974401
969969
965308
967311
973248
980395
985423
986936
984474
978742
972619
967266
966065
970884
975430
981982
987044
987208
983306
975902
971977
968539
966659
968854
974234
981666
986803
987179
983746
978296
972584
967244
966346
967446
973802
979954
985608
988824
986071
979502
972003
967497
966576
969039
974160
979846
985912
988326
985194
979093
973821
968811
963373
965117
971069
979338
985668
987556
983803
978848
972184
968178
967428
968521
974801
981674
986154
986708
985106
980249

The schematic of ADS1291 is as follows:The positive electrode of channel 1 and the electrode driven by the right leg are in contact with the wrist of the left hand; the electrode of the negative electrode of channel 1 is in contact with the electrode pad with the index finger of the right hand when measuring.

The settings of 11 registers (0x01~0x0B) are as follows: 0x02-0xe0-0xf0-0x00-0x81-0x20-0x03-0x00-0x02-0x07-0x0c,Tested with a multimeter, DVDD, AVDD, voltage is normal, lead break detection can also be detected, and 6 bytes of ECG data is acquired using the continuous reading data mode.Use battery power。My physical map is as follows, add some DuPont lines for download and serial print ECG data, only after the data is printed after the collection, when the collection of data is not wired to the outside.

1.What should I do by the above description of me?

2.How to verify that the right leg driver is connected and function?

3.Please give me some suggestions ! The signal of the simulator can be collected. The signal of the human body should be similar, but the normal ECG signal cannot be acquired. I tried to adjust the Lead-off current magnitude and Channel 1 PGA gain.

4.The datasheet has been read many times, and I have noticed many details. For example, the read data is complemented and the SPI clock is set to 1MHZ. The ADS1291 uses an internal clock of 512KHz.,Can you outline which details need attention?

think you . Looking forward to your reply. Shenzhen, China!

  • Hi Yihui,

    Thanks for your post.

    The data you recorded from the ECG simulator looks very good! I suspect that your main problem here is the large contact impedance between your electrodes and the surface of your skin. Try using electrodes with more surface area and add moisturizer/gel to reduce the impedance as much as possible.

    The RLD can be measured directly by setting CH1SET[3:0] to 0010. The output of Channel 1 will become (RLDOUT - RLDREF) times PGA gain.

    1 MHz SCLK is easily fast enough to read the STATUS word + Channel 1 data (48 bits) between each /DRDY falling edge (i.e. 1 / 500 SPS = 2 ms).


    Best Regards,

  • Thanks for your post. I still have 2 questions:

    1. ADS1291 powered from  unipolar supplies(DVDD =3.3V,AVDD = 5V). Please see my schematic diagram。

     What is the value of  70% in the red arrow in the figure below?  and What is the value of  30% in the red arrow in the figure below?  

    70% for unipolar  5-V supply (+5V/-0 V) would set the positive threshold at 0.70 * 5 V + (0 V) = 3.5 V.

    30% for unipolar  5-V supply (+5V/-0 V) would set the negative threshold at 0.30 * 5 V + (0 V) = 1.5 V.

    Is the above calculation correct?

    2.I can now collect ECG signals from the human body. However, the signals of some people who are thin are not very regular(Picture 2), and the signals of fat people are very regular(Picture 1). Will the parameters of the registers need to be adjusted when acquiring data? If necessary, what register parameters need to be adjusted? After setting the register parameters, does it not need to adjust?

    Picture 1

    Picture 2

    thanks.

  • Hello Yihui,

    1. Yes, your calculations for the lead-off comparator thresholds are correct.
    2. The skin-electrode impedance can vary a lot for different people, depending on age, race, gender, overall health, etc. You will have to improve the electrode contact quality when you cannot measure a clean ECG signal. The ADS1291 register settings should not need to change.

    Best Regards,
  • thanks, Ryaan Andrews.