Dear TI Experts,
As discussed in another link:https://e2e.ti.com/support/data_converters/precision_data_converters/f/73/p/693255/2611200#2611200
I am now using ADS1298IPAGR for ECG projects~And now we need to pass the test according to YY1139-2013。But now we fail the test,and really really need your help:
Test condition/Method:
- Condition: Connect a parallel RC network 620k || 4.7nF in series with tested electrode,The idea is to compare the voltage level with and without series impedance. The difference must be less then 20% to pass the test.The test signal frequency 0.5H/10Hz/20Hz/100Hz.~~Here is the test circuit:
- Method:①ECG diagnosis equipment powers on,set the gain to 10mm/mV ②close S1&S2,set S3 to a, those used leads connet to P1 and P2, unused leads connect to P6,then regulate the sinusoidal signal generator to a 1Hz,20mm(peak-to-valley value) signal ③then disconnect S1,measure the change of amplitude.In the stable state,the change must be less then 20%. ④ Then use 0.5Hz,10Hz,20Hz,100Hz,and repeat step ①&②, The output amplitude's difference( disconnect S1 compared with connect S1) should be less then 20%.
Test results:
illustration:1. Sample rate is configured by the register of ADC1_ADS1298CONFIG1