hi
I have tested twice ADS1299 with following configurations and results:
1 test
TEST WITH INTERNAL SIGNALS 1X -(VREFP-VREFN)/2400:-4.5/2400=-0.001875=-1.875mV=-1875uV
CONFIG1=0x96;//250 SPS
CONFIG2=0xD3;//INTERNAL SIGNAL at DC
CONFIG3=0xEC;//INTERNAL REFERENCE BUFFER
CHnSET=0x05;//g=1, input test signal
MISC1=0x00;//SRB1 OPEN
Result:
3 BYTE (MSB first) data Channel 1 for G=1 0x255,0x237,0x218 converting to two-complement
C-2 =-4647 --> Conversion to volts: format Vref=4.5/2 *1/2^23=0.0000002682 V = 0.26822 uV ; -4647*0.26822=-1246uV.
Far from expected -1875uV.
Changing the gain to 2,4,6,8,12,24 the results obtained lack of linearity.
2º Test.
CH1 to CH8 sorted with common negative input connected to pin SRB1 connected to AVSS,AGND.
CH1 to CH8 connected to multiturn potentiometer.
By changing slowly input voltage from 4mV to 8 mV the values obtained are much smaller than 4000 or 8000 uV.
By turning pot. up and down the values also change up and down but the resolution is low.
Any hint about this incorrect behavior.
Thanks