ADS1283: Pulse sync timing; phase delay between input signal and recorded samples.

Part Number: ADS1283

Hi,

We are using the ADS1283 in a data acquisition system that is synchronized to a GPS disciplined rubidium oscillator. The ADS1283 is synchronized on the leading edge of the GPS 1 PPS signal using ADC in pulse sync mode. We apply the sync pulse tdr time prior to next rising edge of PPS so that the first DRDY signal should hit the 1PPS edge at the correct time.

ADC setup:

-Sampling frequency: 500Hz

-ADC clock 4.096MHz

-Filter: SINC+FIR, linear phase,(constant group delay 31 samples)

tdr=62,98046875/500)+(468/4096000)= 126.075ms

The sync pulse is synchronized with the adc clock so that we avoid hitting the “bad spot” tCSDL (±30ns adc clk rising edge window) mentioned in the datasheet.

The falling edge of the DRDY signal hits the 1PPS every 500 samples as expected, confirmed by scope measurement.

So far so good.

We tested the timing by applying a 1ms wide pulse on the ADC input centered 500ms after the GPS 1PPS trigger.

We assumed that we would see the center of the pulse exactly on sample no 1562 after the 1PPS rising edge on our recorded data.

 

A time series plot of the recorded samples shows that the pulse centre was between 1562 ms and 1564 ms and closer to 1562. Analysing the phase of the pulse shows that the pulse centre was 1562.448 ms. We understand that input circuitry does have certain impact on the added delay but 447us is much.

 

Is there any reason why our 1ms pulse center do not hit the 1562 sample but measure an extra phase delay of 447us?

Is there another simple way to test the sync timing?

 

Cheers,

Terje

  • Hi Terje,

    Do you have an RC filter on your inputs (or any other components with significant capacitance) that might result in an RC step response? If we say the input RC filter is mostly charged after 5 time constants, then perhaps there is an RC filter with a time constant of about 90 us (447 us / 5), which wouldn't require a very large filter... a 1kOhm resistor and 90 nF capacitor would be sufficient to introduce this delay.

    Additionally, I would recommend starting with a step input (or increasing the pulse width significantly to approximate a step input) instead of a pulse. I'm not a digital filter expert, but I suspect that the filter response may be showing some combination of both pulse edges in the output, and perhaps testing with a step input would result in an output plot that more closely resembles Figure 41 (The FIR Step Response plot) from the datasheet.