Dear experts,
our customer used ADS1291 pass the FDA certification,
there are some questions:
S1 closed it's ok, S1 open the ECG signal frequency*10,
if you have some suggestion?
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Dear experts,
our customer used ADS1291 pass the FDA certification,
there are some questions:
S1 closed it's ok, S1 open the ECG signal frequency*10,
if you have some suggestion?
Hi Kane,
Thanks for your post.
This test is designed to validate the Recorder system input impedance. I believe the passing requirement is less than 5% signal loss when S1 is open. The results you are seeing ("ECG signal frequency*10") is likely just noise, but it's difficult to say for sure without seeing a scope capture of the ADC input or output data.
Best regards,
Hi Kane,
I can send you an email offline if you need to share proprietary information, but for now I think we can continue this conversation on the forum. The problem described in the initial post is not adequate to provide much more suggestions.
Regards,