Part Number: ADS1298
hi,
we are using ADS1298IPAG on an ECG and I have one question related to input protection.
The ECG shall withstand defibrillator pulse, which will cause about 10V / 30ms on the input protection.
This voltage could appear on ADS1298 analog inputs, through 470Ohm resistor.
The ADS1298 is powered by 3.0V, so the resulting current is about 14mA on each input.
How can I understand Absolute maximum rating for current, is it for each input or for all inputs together? How long is momentary?
Is 14mA+tols x # of inputs supported?
thansk a lot in advance
KR
Vincenzo