ADS1298ECGFE-PDK: 12 Channel ECG system reference design (sbau171d.pdf)

Part Number: ADS1298ECGFE-PDK
Other Parts Discussed in Thread: ADS1298

In the reference design document (sbau171d.pdf)of ads1298 there are some area denoted as NI (Not implemented). So my question is that component are required for working of 12 Channel ECG system ?

The system is working without the NI part ?

  • Hi Koustubh,

    Thanks for your post and welcome to the forum! 

    The system works fine without the NI parts. We left a lot of not implemented components on the board so that customers can easily customize the board. There are many ways to implement an ECG system and depending on the exact application, it may be helpful to try different things. 

    Please see the BIOFAQ in my signature below. There is a video that covers how to implement a 12-lead ECG system.