Good day.
When testing the FDC2214 EVM, I had problems measuring capacitance in the range of 30 ... 2000 pF. Schemes for an asymmetric configuration.
There were two problems:
1) The measured capacity was about 2 times less. But I sort of figured out this problem (https://e2e.ti.com/support/sensors/f/1023/p/489881/2108249?tisearch=e2e-sitesearch&keymatch=FDC2214#2108249).
2) At the minimum measured capacitance (30 pF), I set the parameter Idrive so that the signal amplitude on the LC is no higher than 1.8 V.
After an increase in the measured capacitance (the pictures are shown below), the signal amplitude decreases and when the capacitance exceeds 300 pF, the generation practically ceases.
Cmin (~30 pF)
+100 pF
+100 pF
+300 pF
+300 pF
+1000 pF
+1000 pF
Please tell me whether I made the settings correctly or if I need other values C and L for my task.
Thanks in advance.