Part Number: ADS1292

Dear TI experts:

               I have a project that I need use  ADS1292 for 1-LEAD ECG applicationwithout  RLD.  The pin assignments in the datasheet of ADS1292  describe RLDIN/RLDREF and RLDINV connect to AVDD if not used,The RLD function  is not enabled in the register,Can I not connect to AVDD,thanks

  • Hi Leo,

    These pins should be connected to AVDD to reduce power consumption in the device. Additionally, if there is an incorrect setting in the registers you may accidentally connect a floating voltage internal to the device which can lead to unexpected issues.