Hello,
we are trying to use the ADS124S08 for measuring temperatures with type K thermocouples.
But currently, we have some problems with the accuracy of our measurements. For testing, we have connected a Beamex MC2 calibrator to analog input 0 and analog input 1 and are generating voltages between -5 and 55mV. But the measured voltages always have an offset of at least 1 - 1.5 percent. The offset also does not seem to be linear (see attached chart). We tried with different settings using different filters, different VBIAS value and chopping, but this produced similar results.
Also, sometimes the voltage jumped unexpectedly, up to 0.05 mV, like noise. This led to jumps in the temperature of around 1-1.2 degree celsius. We could not find a pattern in the noise.
All tests were done under room temperature.
For the test in the chart, we used the following settings:
1. Set the correct inputs in the Input Multiplexer register.
2. Disable the reference monitor and reference buffers, enable the internal reference in the Reference Control register.
3. Enable the PGA and set the PGA gain to 32 in the Gain Setting Register.
4. Set VBIAS to (AVSS + AVDD) / 2 and enable VBIAS on the correct input pin in the Sensor Biasing register.
5. Enable Chopping and sinc filter in the Datarate register.
This is the our schematic for the ADS124S08, we tried to keep it as close to the reference schematics in the datasheet as possible:
This is the schematic for the analog inputs:
So currently we are not sure if the error is caused by a wrong configuration of the converter, if we have an error in our schematic or if we just expect too high precision. Thanks in advance for any help!
Best regards,
Jakob