Hi
I saw a EEG system using ADS1299 that used Mid-supply directly connected to the body instead of connect Biasout. Is this correct and can we design the same way? If so, for Vsuuply=5 should we use 2.5 Volts and 0 Volt for 2.5 Supply?
Thanks
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hi
I saw a EEG system using ADS1299 that used Mid-supply directly connected to the body instead of connect Biasout. Is this correct and can we design the same way? If so, for Vsuuply=5 should we use 2.5 Volts and 0 Volt for 2.5 Supply?
Thanks
Hi Mojtaba,
Am implementation like this is fine.
For power supplies, the device can be run in a unipolar configuration where AVDD = 5V and AVSS = 0V, or in a bipolar configuration where AVDD = 2.5V and AVSS = -2.5V. For additional information, please consult the datasheet.
Hi Alexander
Thanks for your answer
I mean for Vsupply=5 V, can we connect midsupply=2.5 V to the body instead of BIASOUT?
Hi Mojtaba,
You can connect mid-supply, 2.5V, to the body. This will ensure that the measurement is within the common-mode range of the PGA, however you would not get the added benefit of canceling out additional common-mode noise by using the BIASOUT feature. You would have to rely on the common-mode rejection ratio of the PGA amplifiers and digital filtering.