ADS124S08: When I try to read the analogue values, I read always 0.

Part Number: ADS124S08

I work in a project where I have to communicate with two ADS124s08, I am trying now to read the analogue value of one of them, I have connected a square signal between AIN0 (Positive channel) and AIN1 (Negative Channel), AIN1 is connected to GND. The square signals has a low level of +100mV and a high level of +200mV and a frequency of 1Hz. Please find attach a picture of the input measurement with the oscilloscope.

This is our actual HW connection:



These are the steps that I follow to measure the analogue values between AIN0 and AIN1:
- Read AD Status
- Write AD Registers
- Read AD Registers and check that the write/read data are the same
- Send digital signal to start
- I have a external interrupt with falling edge detection connected to DRDY signal
- Read data direct, sending 0 through SPI when the external interrupt is triggered. SPI works with a frequency of 4MHz.

Result:
I receive in the first byte the status as I expected, because I configured it so, but the rest three bytes are always 0.

I have tested with different configurations, but still does not work.

This is my actual Register configuration:

- Inpmux=0x01; (AIN0 as positive channel, AIN1 as negative channel)
- Pga=0x08; Pga enabled, Pga gain = 1
- Datarate= 0x1D; (Global chop disabled, internal clock selected, Continuous conversion mode, Low latency filter, 4000SPS.
- Ref = 0x3A; (Reference monitor disabled), Positive and negative reference buffer disabled, Internal reference selected, Internal reference is always on
- Idacmag=0x00; (Pga rail flag disabled, low side power switch open, idac magnitude off)
- Idacmux=0xDD; (Idac1 and Idac2 disconnected)
- Vbias=0x00; Vbias disconnected
- Sys=0x11; (System monitor disconnected, 8 samples, spi timeout disabled, crc disabled, status enable)
- Ofcal0,Ofcal1,Ofcal2,Fscal0,Fscal1,Fscal2=0;
- Gpiodat=0x00; (GPIOs configured as output, GPIOs low)
- GpioCon=0x00; (GPIOs configured as analog inputs.

I selected PGA enabled and Gain 1 because I saw in datasheet that for Single-ended input and bipolar power supply the Pga should be enabled, but I have also tested with PGA disabled.

I disable the reference buffers because I saw in datasheet that when the internal reference is selected for measurements, these must be disabled.

I have also tested with Sinc3 filter instead Low latency filter.

I have also tested enabling the Reference monitor and the PGA output rail flag to check if there is any fault, but the status that I receive does not change, it is always 0x80, without any fault.

These are some pictures of DRDY, SCLK and DOUT signals:

- DRDY signal and SCLK


- SCLK and DOUT


- SCLK and DOUT with zoom


I have also check the status of the following signals during the measurement:

- RESET is always high
- START is always high
- CS is always low

Please could you tell me if the configuration is correct and if I do something wrong? Because I read always 0.

Thank you very much in advance.

Kind Regards

Jenifer Blanco

  • Hi Jenifer,

    Welcome to the E2E forum!  Are you writing to the OFCAL and FSCAL registers?  You stated that "- Ofcal0,Ofcal1,Ofcal2,Fscal0,Fscal1,Fscal2=0;" The FSCAL2 register should not be 0, but rather 0x40.  If this value is 0, then the conversion result will be incorrect and could be 0.  See Figure 81 on page 56 of the ADS124S08 datasheet.

    One other thing I noticed in the schematic is you are missing the required 330nF cap between AVDD and AVSS.

    Best regards,

    Bob B

  • Hi Bob,

    thank you very much for the quick answer, I did not realized about FSCAL register. About the schematic you are right this is something we need to check. I will let you know if with this change I can read properly the values. 

    Thank you ver much again

    Best regards

    Jenifer Blanco

  • Hi Bob, 

    thank you very much, I have already test it and works perfect. 

    Best Regards

    Jenifer Blanco