Part Number: ADS114S08
I am using the ADS114S08B in single-end inputs, 0-5V range with reference=0V and full scale=5V, with PGA bypassed.
The negative input is AINCOM, while the positive input is successively connected to AIN0 to AIN11, each one with a duration of 1 ms.
With the scope, I see that each time an input is selected, an input current is flowing to it: the ADC is sourcing about 150 µA.
According to the datasheet, the input current should be "0.5 nA typ.", for input voltages between 0.1 and 4.9V. Some input voltages are between 0 and 0.1V, but this current exists even when the voltage is 0.3V. Input signal AIN11 is +5V (signal "-15V" is disconnected), and I am afraid AIN0 might be >5V.
I have checked that:
- PGA is bypassed
- positive and negative reference buffer bypassed are disabled
- IDAC (excitation current sources) are disabled and disconnected
- VBIAS is disconnected (which implies disabled)
- burn-out current sources are disabled
The current really seems to come from the PGA or the core ADC: in register SYS, if I set SYS_MON=001 (disconnect PGA inputs from multiplexer and short them), the input current disappears (but the ADC gives a zero value).
This input current makes the converted values imprecise, when they are not low-impedance: this is the case for AIN0 (2k), AIN3 (>10k), AIN4 (2k), AIN8 (ISENSE_ANA have 1k impedance).
What can be the reason and is there a solution?