Hello,
I have to design a 0.01 degrees accuracy PT100 measurement. Therefore I was looking for the ADS1247 or ADS1248 in a 3 or 4 wire RTD measurements.
I just read a few threads, the sbaa110 and sbaa046 application note.
Do you have a spreadsheet to fill in all the errors of the ADS to calculate easily the effective accuracy of the measurement?
Our range is from 60 to 200 Ohms, I choose a current of 1mA, Rbias of 1200 Ohm equals 2,4V reference and a PGA Gain of 16.
This results in a 90mV voltage swing when the Rcomp is chosen to 110 Ohms for 25 degrees.
What would be the drift of the measurement in a temperaturrange from 0 to 60 degrees.
Is a two point Gain and Offset calibration sufficient to reach the accuracy? Of course I know typical PT100 elements even in Type A class have a much bigger tolerance, so I have to calibrate the PT100 sensor as well.
Do I have do calibrate Gain + Offset on 0 and 60 degrees for sufficient accuracy?
Thank you and best regards