Dear technical team,
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Dear technical team,
Hello Yufei,
What is the orientation of the incident light relative to the normal of the DMD?
What is the wavelength of light you are using?
What is meant by the "first order"?
Please be aware that there are only two operational states for a DMD - "on" or "off". Parked ("flat) is not an operational state.. During operation, the mirrors may briefly pass through flat, but do not remain there.
Fizix
Hi Fizix,
Many thanks for your prompt reply. We are using a 532nm laser. The angle of incident light relative to the normal of the DMD is 12 degree.
Sorry for the confusion, we are capturing the first diffraction order of the reflected pattern.
I have attached a schematic figure for better understanding on the issue. It seems that when DMD switching the patterns, all micromirrors will be reset to 'On' state. It will cause a power boost to our detector.
regards,
Yufei
Hello Yufei,
First, your diagram is very good and very helpful.
Please see Using Lasers with DLP® DMD technology By my calculations, you are operating somewhere near the 4,4 order.
I would suggest trying the incident light at about 22.5 degrees from the normal. This will operate near the 7,7 order normal to the DMD. This order should be much brighter, and the "anomaly" will be much lower relative to the signal.
To determine which order your detector is at, turn the DMD off and find the brightest order in the output field. This is the 0,0 order. Count along the diagonal of the diffraction pattern (refer to the linked application note above). This will be along the n=m line of the orders - i.e. (n,n).
I believe what is happening is that at the order you are working in, the pixels that are transitioning from "on" to "off" briefly pass through a blaze condition. The momentary blaze condition is MUCH more efficient than the order you are using so that you get a spike in the signal.
Converting to the illumination scheme I suggested, should correct what you are experiencing.
Fizix