Hi,
I would like to know if there is a way of calculating the spectral resolution through the use of an equation-based formula.
Thanks.
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hi,
I would like to know if there is a way of calculating the spectral resolution through the use of an equation-based formula.
Thanks.
Hello again Danson,
At this time the number we have published is conservative based on measurements made with actual units. We are continuing to make more measurements. At some point in the future we will have a note on design considerations. Until then we do not have a formula that we publish.
We certainly welcome people to report back anything that they are able to measure regarding resolution.
Fizix
David,
So the resolution of the system is a convolution of slit width, optical transfer function of the system, and DMD resolution. So for a given optics and DMD resolution, the smaller slit does provide finer resolution, but less light enters the system, which increases SNR. As far as slit widths, the limited data we have captured suggests:
As you can see from this trend, 3nm resolution would require an extremely small slit. For this application, the optics would need to be modified for finer resolution.
Best regards,
Pedro
David,
We have published SNR and wavelength resolution data in terms of slit with and DMD pattern width in this 2016 SPIE Photonics West paper: DLP NIRscan Nano: an ultra-mobile DLP-based near-infrared Bluetooth spectrometer.
The NIRscan Nano is using G12180-010A detector from Hamamatsu. Hamamatsu also offers 1-stage cooled G8605-11 and a two-sate cooled G8605-21. The cooled detectors decreases the dark current and achieves a higher detectivity (D*). Note that the NIRscan Nano electronics does not include a TEC cooler, the cooled detectors have different pinout, and the detector is glued to the chassis. Replacing the detector requires the system to be re-calibrated.
Best regards,