Hello,
I am working on BQ20z80. If I change the data Golden Pack to a calibrate battery, does that delete the calibration values (current, tension, temperature, coulombe count) ?
Regards
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hello,
I am working on BQ20z80. If I change the data Golden Pack to a calibrate battery, does that delete the calibration values (current, tension, temperature, coulombe count) ?
Regards
Hi Dimitri,
I don't not quite understand your question.
Whenever you run calibration, the cal values in data flash will be updated.
Hi Jackie
My circuit is calibrate, I attach my cells and then if I put golden data into that battery, does it change the calibrate data?
My circuit arrive calibrate in temperatrue current and tension. I attach my cells then I put data image (.dif) coming from a golden pack using the same cells lot.
I think that, send data image will change temperature, current and tension calibration ? am I right ? is it possible to send data image without calibration change ?
thank you for your response
The calibration data will be updated to whatever values stored in the DFI whenever you program the device with a dfi file.
You can:
1. read out a gg file before programing a dfi file (eg. file1.gg)
2. program dfi
3. read out a gg file after dfi update (eg. file2.gg)
4. Replace the calibration data in file2.gg with file1.gg
5. Program file2.gg to the device