BQ24133 battery detection test

Title: Battery Detection Test

-       Whenever the EVM behave from non-charging mode to charging mode would mean the battery detection flow chart is applied.

-       8mA discharge current is a constant current.

 Test Setup:

1) Set power supply to 12V, 3A.

2) Set DMM to DCI, 10A Range, Fast Mode.

3) Set E-Load to Resistance function with 8 ohm resistance.

4) Connect DMM to PC by USB.

5) Connect the entire instruments as show in Figure 1 below.

 

Test Sequence:

1) Ensure EVM is under non-charging mode, Current reading is 0A.

2) Start collecting reading from DMM using SCPI “READ?”

3) Close E-load to allow EVM start charging.

4) Stop Data collection from PC.

 Result:

 

Expected   Reading

Result

Before start

0A

0A

First 1s

-8mA

112mA

End

125mA

112mA

 

Discussion:

1) Base on the flow chart, EVM will first apply 8mA discharge current for 1s. Thus the DMM should read a -8mA   at first.

2) The DMM have a spec to read 19 data per second, thus we can ensure the DMM performance is good enough to catch the -8mA within the first second of current flow.

 

Conclusion:

-      The EVM does not apply the battery detection flow chart as the 8mA discharge current was not detected.

-       The flow chart is invalid.

Question: what happen if a fully charged battery is detected?