Dear All,
This is Bhairu from Cardiac Design Lab.
We have two design ALPHA & POC. We have used BQ24192 in both the design. The battery is not getting charged when device is ON, However same is getting charged when device is OFF. We carried out the same experiment with Alpha design & same battery, battery is getting charged even though device is ON or OFF. The Bq24192 schematic and testing results are as below.
Fig: BQ24192 Schmatic
LOAD status | Charge Status | Design | Adapter voltage(V) |
I/P current | Load current | Battery current(A) |
Charging current(A) |
Disharging current(A) |
Remarks |
LOAD ON | Charge | ALPHA | 5 | 1.66 | 1.576998 | 0.083002 | 0.083002 | I/P Current = Charging current + Load current Here battery is charging |
|
9 | 0.98 | 0.898604 | 0.081396 | 0.081396 | |||||
12 | 0.76 | 0.682531 | 0.077469 | 0.077469 | |||||
POC | 5 | 1.06 | 1.126426 | 0.066426 | 0.066426 | Load current = Input current + Battery discharge current here battery is not charging. |
|||
9 | 0.63 | 0.644458 | 0.014458 | 0.014458 | |||||
12 | 0.48 | 0.501598 | 0.021598 | 0.021598 | |||||
LOAD OFF | Charge | ALPHA | 5 | 0.59 | 0.717567568 | ||||
9 | 0.35 | 0.766216216 | |||||||
12 | 0.27 | 0.788108108 | |||||||
POC | 5 | 1.15 | 1.398648649 | ||||||
9 | 0.67 | 1.466756757 | |||||||
12 | 0.52 | 1.517837838 |
Table: The testing results
So anyone suggest me what could be the issue is. Please reply me at the earliest.
Thanks and Regards,
Bhairu