PGA970: Generated frequency is half of the expected frequency
Part Number: PGA970
I am using the PGA970EVM and am facing issues with the waveform generator and bandpass filter. The generated frequency is half of what is set in the waveform generation settings (WFM).
In the below example I have configured the WFM Frequency to 1kHz (sine wave). When measured at pins TP27 and TP29, i.e. S1P and S1N, the frequency appears to be 500Hz.
The PGA970 firmware is configured via the GUI with the following settings:
From the bandpass filter settings, the bandpass filter is able to let through the signal with the center frequency set at 1000Hz (250Hz Band Width) - even through the wave is measured at 500Hz, and according to the results on ADC1, amplitude is demodulated correctly (note that I have tied the output of the generated wave directly to J3 (ADC1) for testing).
I have also replicated this with a function generator, generating a 1kHz sine wave. When the bandpass filter is configured with a center frequency of 1000Hz, I am unable to get the demodulated amplitude but when configured with a center frequency of 2000Hz, the amplitude can be measured.
Why could this be happening?
  • Hi Ric,
    Can you try 2 things for me?
    1. Read the waveform RAM from the waveform ram page and see if they match the table on the ADC settings page
    2. Change the amplitude to 295mV, repopulate the table and try again
    -Cole
  • Hi Cole,
    The following information was collected by directly feeding a non-externally amplified waveform by altering the jumpers on J2 and J7 and skipping the ADC filtering by having J1 and J4 jumpers in place. I can find no correlation between the waveform data table on the ADC settings page and the Data and Waveform RAM page. Please refer to the following screenshots and files:
    300mV, 1kHz waveform configuation: (Please note the ADC1 Data reflects very close to 300mV)
    Data and waveform RAM programming table for the 300mV, 1kHz waveform configuration:
    300mV, 1kHz Waveform Data table file from the table under 'waveform generation' on the ADC settings page:
    Address(Hex)	MSB(Hex)	LSB(Hex)
    0	00	0C
    2	00	25
    4	00	3E
    6	00	56
    8	00	6F
    A	00	88
    C	00	A1
    E	00	B9
    10	00	D2
    12	00	EB
    14	01	03
    16	01	1C
    18	01	35
    1A	01	4D
    1C	01	66
    1E	01	7E
    20	01	97
    22	01	AF
    24	01	C8
    26	01	E1
    28	01	F9
    2A	02	12
    2C	02	2A
    2E	02	42
    30	02	5B
    32	02	73
    34	02	8C
    36	02	A4
    38	02	BC
    3A	02	D5
    3C	02	ED
    3E	03	05
    40	03	1D
    42	03	36
    44	03	4E
    46	03	66
    48	03	7E
    4A	03	96
    4C	03	AE
    4E	03	C6
    50	03	DE
    52	03	F6
    54	04	0E
    56	04	25
    58	04	3D
    5A	04	55
    5C	04	6D
    5E	04	84
    60	04	9C
    62	04	B3
    64	04	CB
    66	04	E2
    68	04	FA
    6A	05	11
    6C	05	28
    6E	05	40
    70	05	57
    72	05	6E
    74	05	85
    76	05	9C
    78	05	B3
    7A	05	CA
    7C	05	E1
    7E	05	F8
    80	06	0E
    82	06	25
    84	06	3C
    86	06	52
    88	06	69
    8A	06	7F
    8C	06	95
    8E	06	AC
    90	06	C2
    92	06	D8
    94	06	EE
    96	07	04
    98	07	1A
    9A	07	30
    9C	07	46
    9E	07	5B
    A0	07	71
    A2	07	87
    A4	07	9C
    A6	07	B2
    A8	07	C7
    AA	07	DC
    AC	07	F1
    AE	08	07
    B0	08	1C
    B2	08	31
    B4	08	45
    B6	08	5A
    B8	08	6F
    BA	08	83
    BC	08	98
    BE	08	AC
    C0	08	C1
    C2	08	D5
    C4	08	E9
    C6	08	FD
    C8	09	11
    CA	09	25
    CC	09	39
    CE	09	4D
    D0	09	60
    D2	09	74
    D4	09	87
    D6	09	9B
    D8	09	AE
    DA	09	C1
    DC	09	D4
    DE	09	E7
    E0	09	FA
    E2	0A	0C
    E4	0A	1F
    E6	0A	32
    E8	0A	44
    EA	0A	56
    EC	0A	69
    EE	0A	7B
    F0	0A	8D
    F2	0A	9F
    F4	0A	B0
    F6	0A	C2
    F8	0A	D4
    FA	0A	E5
    FC	0A	F7
    FE	0B	08
    100	0B	19
    102	0B	2A
    104	0B	3B
    106	0B	4C
    108	0B	5C
    10A	0B	6D
    10C	0B	7D
    10E	0B	8E
    110	0B	9E
    112	0B	AE
    114	0B	BE
    116	0B	CE
    118	0B	DE
    11A	0B	ED
    11C	0B	FD
    11E	0C	0C
    120	0C	1B
    122	0C	2B
    124	0C	3A
    126	0C	48
    128	0C	57
    12A	0C	66
    12C	0C	74
    12E	0C	83
    130	0C	91
    132	0C	9F
    134	0C	AD
    136	0C	BB
    138	0C	C9
    13A	0C	D7
    13C	0C	E4
    13E	0C	F1
    140	0C	FF
    142	0D	0C
    144	0D	19
    146	0D	26
    148	0D	32
    14A	0D	3F
    14C	0D	4B
    14E	0D	58
    150	0D	64
    152	0D	70
    154	0D	7C
    156	0D	87
    158	0D	93
    15A	0D	9F
    15C	0D	AA
    15E	0D	B5
    160	0D	C0
    162	0D	CB
    164	0D	D6
    166	0D	E1
    168	0D	EB
    16A	0D	F6
    16C	0E	00
    16E	0E	0A
    170	0E	14
    172	0E	1E
    174	0E	27
    176	0E	31
    178	0E	3A
    17A	0E	43
    17C	0E	4D
    17E	0E	56
    180	0E	5E
    182	0E	67
    184	0E	6F
    186	0E	78
    188	0E	80
    18A	0E	88
    18C	0E	90
    18E	0E	98
    190	0E	A0
    192	0E	A7
    194	0E	AE
    196	0E	B6
    198	0E	BD
    19A	0E	C3
    19C	0E	CA
    19E	0E	D1
    1A0	0E	D7
    1A2	0E	DE
    1A4	0E	E4
    1A6	0E	EA
    1A8	0E	F0
    1AA	0E	F5
    1AC	0E	FB
    1AE	0F	00
    1B0	0F	05
    1B2	0F	0A
    1B4	0F	0F
    1B6	0F	14
    1B8	0F	19
    1BA	0F	1D
    1BC	0F	22
    1BE	0F	26
    1C0	0F	2A
    1C2	0F	2E
    1C4	0F	31
    1C6	0F	35
    1C8	0F	38
    1CA	0F	3C
    1CC	0F	3F
    1CE	0F	42
    1D0	0F	44
    1D2	0F	47
    1D4	0F	4A
    1D6	0F	4C
    1D8	0F	4E
    1DA	0F	50
    1DC	0F	52
    1DE	0F	54
    1E0	0F	55
    1E2	0F	57
    1E4	0F	58
    1E6	0F	59
    1E8	0F	5A
    1EA	0F	5B
    1EC	0F	5B
    1EE	0F	5C
    1F0	0F	5C
    1F2	0F	5C
    295mV, 1kHz waveform configuation: (Please note the ADC1 Data reflects very close to 295mV)
    Data and waveform RAM programming table for the 295mV, 1kHz waveform configuration: (Nothing appears to have changed).
    Note: If I power off, then on again, this table looks completely different every time, even when reconfigured with identical settings:
    295mV, 1kHz Waveform Data table file from the table under 'waveform generation' on the ADC settings page:
    0	00	0C
    2	00	24
    4	00	3D
    6	00	55
    8	00	6D
    A	00	86
    C	00	9E
    E	00	B6
    10	00	CE
    12	00	E7
    14	00	FF
    16	01	17
    18	01	2F
    1A	01	48
    1C	01	60
    1E	01	78
    20	01	90
    22	01	A8
    24	01	C0
    26	01	D9
    28	01	F1
    2A	02	09
    2C	02	21
    2E	02	39
    30	02	51
    32	02	69
    34	02	81
    36	02	99
    38	02	B1
    3A	02	C9
    3C	02	E0
    3E	02	F8
    40	03	10
    42	03	28
    44	03	40
    46	03	57
    48	03	6F
    4A	03	87
    4C	03	9E
    4E	03	B6
    50	03	CD
    52	03	E5
    54	03	FC
    56	04	14
    58	04	2B
    5A	04	42
    5C	04	5A
    5E	04	71
    60	04	88
    62	04	9F
    64	04	B6
    66	04	CD
    68	04	E4
    6A	04	FB
    6C	05	12
    6E	05	29
    70	05	40
    72	05	57
    74	05	6D
    76	05	84
    78	05	9B
    7A	05	B1
    7C	05	C8
    7E	05	DE
    80	05	F4
    82	06	0B
    84	06	21
    86	06	37
    88	06	4D
    8A	06	63
    8C	06	79
    8E	06	8F
    90	06	A5
    92	06	BB
    94	06	D1
    96	06	E6
    98	06	FC
    9A	07	11
    9C	07	27
    9E	07	3C
    A0	07	51
    A2	07	67
    A4	07	7C
    A6	07	91
    A8	07	A6
    AA	07	BB
    AC	07	D0
    AE	07	E4
    B0	07	F9
    B2	08	0E
    B4	08	22
    B6	08	37
    B8	08	4B
    BA	08	5F
    BC	08	73
    BE	08	87
    C0	08	9B
    C2	08	AF
    C4	08	C3
    C6	08	D7
    C8	08	EB
    CA	08	FE
    CC	09	12
    CE	09	25
    D0	09	38
    D2	09	4B
    D4	09	5F
    D6	09	72
    D8	09	84
    DA	09	97
    DC	09	AA
    DE	09	BD
    E0	09	CF
    E2	09	E2
    E4	09	F4
    E6	0A	06
    E8	0A	18
    EA	0A	2A
    EC	0A	3C
    EE	0A	4E
    F0	0A	60
    F2	0A	71
    F4	0A	83
    F6	0A	94
    F8	0A	A6
    FA	0A	B7
    FC	0A	C8
    FE	0A	D9
    100	0A	EA
    102	0A	FA
    104	0B	0B
    106	0B	1B
    108	0B	2C
    10A	0B	3C
    10C	0B	4C
    10E	0B	5C
    110	0B	6C
    112	0B	7C
    114	0B	8C
    116	0B	9C
    118	0B	AB
    11A	0B	BA
    11C	0B	CA
    11E	0B	D9
    120	0B	E8
    122	0B	F7
    124	0C	05
    126	0C	14
    128	0C	23
    12A	0C	31
    12C	0C	3F
    12E	0C	4D
    130	0C	5B
    132	0C	69
    134	0C	77
    136	0C	85
    138	0C	92
    13A	0C	A0
    13C	0C	AD
    13E	0C	BA
    140	0C	C7
    142	0C	D4
    144	0C	E1
    146	0C	ED
    148	0C	FA
    14A	0D	06
    14C	0D	13
    14E	0D	1F
    150	0D	2B
    152	0D	36
    154	0D	42
    156	0D	4E
    158	0D	59
    15A	0D	64
    15C	0D	70
    15E	0D	7B
    160	0D	86
    162	0D	90
    164	0D	9B
    166	0D	A5
    168	0D	B0
    16A	0D	BA
    16C	0D	C4
    16E	0D	CE
    170	0D	D8
    172	0D	E1
    174	0D	EB
    176	0D	F4
    178	0D	FE
    17A	0E	07
    17C	0E	10
    17E	0E	18
    180	0E	21
    182	0E	2A
    184	0E	32
    186	0E	3A
    188	0E	42
    18A	0E	4A
    18C	0E	52
    18E	0E	5A
    190	0E	61
    192	0E	68
    194	0E	70
    196	0E	77
    198	0E	7E
    19A	0E	84
    19C	0E	8B
    19E	0E	92
    1A0	0E	98
    1A2	0E	9E
    1A4	0E	A4
    1A6	0E	AA
    1A8	0E	B0
    1AA	0E	B5
    1AC	0E	BB
    1AE	0E	C0
    1B0	0E	C5
    1B2	0E	CA
    1B4	0E	CF
    1B6	0E	D4
    1B8	0E	D8
    1BA	0E	DD
    1BC	0E	E1
    1BE	0E	E5
    1C0	0E	E9
    1C2	0E	ED
    1C4	0E	F1
    1C6	0E	F4
    1C8	0E	F7
    1CA	0E	FB
    1CC	0E	FE
    1CE	0F	01
    1D0	0F	03
    1D2	0F	06
    1D4	0F	08
    1D6	0F	0B
    1D8	0F	0D
    1DA	0F	0F
    1DC	0F	11
    1DE	0F	12
    1E0	0F	14
    1E2	0F	15
    1E4	0F	16
    1E6	0F	17
    1E8	0F	18
    1EA	0F	19
    1EC	0F	1A
    1EE	0F	1A
    1F0	0F	1A
    1F2	0F	1B
  • Hi Ric,
    I forgot to give an important step, sorry about that. Please change the page to 0x1 and then capture the screenshot. You can also right click the center table and choose the export the data and put it in a code format like you did with the waveform data.
    See the 0x01 page address for waveform ram below.
    How what the output waveform though. Did the frequency still stay around 500Hz?
    -Cole
  • Hi Cole,
    I have rerun these configurations for page 0x01. No differences on page 0x01 were noticed between 300mV and 295mV configurations. The waveform is still producing 500Hz frequency.
    Data and waveform RAM programming table for the 295mV, 1kHz waveform configuration, page 0x01:
    Page Address: 	1
    Page	Word0	Word1	Word2	Word3	Word4	Word5	Word6	Word7
    0	F490	FD45	61A6	8AF4	17C5	21AF	8AAF	35EE
    1	B2CE	ED1D	7A49	08AA	09FF	5A29	0B7E	E775
    2	6695	76CE	FCF7	975E	6A48	F68D	0484	D52D
    3	165C	8C3D	6F9A	8E47	1EF6	FB8B	6F93	C68F
    4	355A	6BF9	BD5C	0577	9229	5817	2411	D658
    5	6A6A	B668	5746	8540	3F05	9D59	6CB3	F8C5
    6	0D9F	AFD4	FC5E	0077	9CAE	7DC4	46FF	F7C1
    7	EF73	CE30	2F31	014B	FF54	366F	470B	29D8
    8	90C6	FCF8	FA08	1CE5	D8EC	0B47	FCB5	4753
    9	FA23	1134	2743	850A	5B72	8289	D71C	2180
    A	F6CF	D69A	3745	D243	AFCF	4C6A	0BAC	0F11
    B	7B3E	4A37	E577	B44E	7FA1	A308	0B81	183A
    C	4A6D	C65A	44F0	5AAF	2E9D	2CE3	1F24	3229
    D	7BC2	BCA3	65DF	137F	5308	10D3	ACB3	226A
    E	2AFF	CEA7	6862	F7B6	3EE9	AD8D	C186	43EC
    F	5F98	FD82	25AF	2C05	7828	5FCF	EA3F	624B
    Data and waveform RAM programming table for the 300mV, 1kHz waveform configuration, page 0x01:
    Page Address: 	1
    Also, I have included data for 0x08 since it seems possible that the waveform RAM is stored there according to the memory map. However, I have noticed that the data on page 0x08 changes every time I regenerate the same waveform.
    Data and waveform RAM programming table for the 295mV, 1kHz waveform configuration, page 0x08:
    Page Address: 	8
    0	8804	4204	6C0B	010F	700D	9107	CE00	2509
    1	E002	DE05	770C	1B0F	920C	DE05	E002	940A
    2	C40D	0E08	3D00	8708	FE0D	D40E	C80A	2803
    3	7906	0902	CF09	9E0E	520E	4B09	FF00	9107
    4	C805	9902	600A	C00E	210E	D708	6D00	0E08
    5	8B0E	5A0E	3809	4801	5107	7B0D	080F	BA0B
    6	8804	BA0B	080F	4E0D	FC06	7801	5F09	680E
    7	B00D	F40E	2C0B	B603	E404	140C	140F	060D
    8	D007	BA0D	F10E	1B0B	9E03	2905	3F0C	190F
    9	7E0E	680E	5F09	7801	FC06	360D	0D0F	F70B
    A	0B0B	5703	4005	4D0C	1A0F	BA0C	F405	F802
    B	7E0E	9709	2F01	6707	860D	FB0E	2C0B	B603
    C	7C07	7B0D	FE0E	9C0B	2B04	7104	AB0B	0B0F
    D	8C0B	060F	590D	6707	2F01	4B09	520E	AA0E
    E	F10E	5C0B	CD03	2905	3F0C	190F	C70C	3706
    F	420E	9E0E	F409	0902	A506	360D	0D0F	D90B
    Data and waveform RAM programming table for the 300mV, 1kHz waveform configuration, page 0x08:
    Page Address: 	8
    0	F107	0A0E	2A0F	2A0B	4E03	9C05	910C	5C0F
    1	3F0F	9F0D	1A07	AF01	FA09	DE0E	800E	4D09
    2	690A	A402	3C06	260D	580F	480C	FA04	2504
    3	F70A	1D0F	000E	0708	A100	2509	6F0E	DE0E
    4	5B02	9F0A	050F	310E	4508	8800	1109	880E
    5	0E06	0503	D40A	1D0F	000E	0708	D200	4D09
    6	690A	4202	EE06	870D	440F	AE0B	DE03	1105
    7	AA0D	3C0F	5C0B	C603	5705	9F0C	5C0F	D70C
    8	D409	D10E	800E	2509	6F00	5A08	270E	0A0F
    9	E105	0C0D	580F	2B0C	6D04	B304	3A0C	5B0F
    A	440F	AE0B	DE03	4005	740C	5C0F	C90C	F805
    B	1D03	2A0B	2A0F	E10D	9C07	4D01	AE09	B60E
    C	380F	9F0D	4607	AF01	FA09	DE0E	6F0E	FD08
    D	580F	0C0D	E105	3603	3B0B	2E0F	D60D	8707
    E	F70A	D502	0E06	0C0D	580F	480C	CB04	8404
    F	F901	0407	930D	420F	5C0B	9603	B305	BB0C
  • Hi Ric,
    Thanks for taking the data. Not the result I was expecting.
    Please give me a couple of days and let me see if I can reproduce you results.