Other Parts Discussed in Thread: HDC3022
When humidity is measured in a constant temperature and humidity chamber, the error becomes larger.
When the HDC3020 is placed in a constant temperature and humidity chamber to measure humidity, the lower the humidity, the larger the error.
*There was almost no error for temperature.
Measurement Procedure:
1) Put HDC3020 and Assmann Psychrometer into the constant temperature and humidity chamber.
2) Change the humidity in the order of 25°C/30->10->30->50->70->90->70->50->30%RH.
Measurement results:
Chamber | 30 | 10 | 30 | 50 | 70 | 90 | 70 | 50 | 30 |
Assmann | 28.9 | 6.5 | 28.5 | 50.7 | 71.8 | 92.8 | 71.1 | 49.7 | 28.3 |
HDC3020 | 32.1 | 9.9 | 31.4 | 53.3 | 73.1 | 93.0 | 73.2 | 53.3 | 31.6 |
Error | 3.2 | 3.4 | 2.9 | 2.6 | 1.3 | 0.2 | 2.1 | 3.6 | 3.3 |
A few days later, when we checked the readings of the HDC3020 and the Assmann Psychrometer in the room, the error was 0.1% at 25°C/60%RH humidity.
Is there anything I should pay attention to when measuring humidity in a constant temperature and humidity chamber?
*We waited 20 minutes after the humidity in the thermostatic chamber reached the set value before recording the measured value.
*The HDC3020 was not housed in a case and the sensor board was left bare for the measurement.