Other Parts Discussed in Thread: HDC2021, HDC3020
Tool/software:
Hello TI,
With refer to last case attended by TI's Sakeenah Khan, The reply unable to send because the Mail delivery system failed.
With respect to same subject the reply is given below.
- Regarding the test setup:
- Does the test chamber have a controlled humidity? - Yes, we have tested 3 Sensors ( Device) in the Controlled Humidity chamber. PFA.
- Do you have any humidity reference measurement besides the 3 HDC3021s? - Yes, Rotronic / HP22-A is considered as Reference device. PFA.
- Are the temp & humidity stable and constant throughout the chamber, or are there potential "hot spots"? - Temp & humidity are not constant throughout the test because we have to take accuracy at full range over the temp for -40 to 65 degrees and Humidity 20 to 95%
- Could you share a picture of the test setup inside of the chamber? Yes, PFA.
- What materials and methods were used during storage, handling and assembly?
- Temperature and humidity conditions? - Ambient Temp and Hum.
- What type of packing materials, and was pink foam used? When we received this Sensor the Pink foam kept as it is on sensor during the soldering Process and as per the datasheet we had kept the Hot air gun temp 260 Degree.
- Have devices been exposed to high/low temperature/humidity for a prolonged time? No.
- Is there any possibility of chemical contamination? E.g. chemicals in epoxies, glues, adhesives may outgas during baking/curing processes and contaminate the sensor; cleaning agents, no-clean flux, or strong blasts from aerosol dusters are also problematic - No.
- If tweezers were used to remove the tape from the HDC3021, is it possible the sensor was scratched? No.
- The Silicon Users Guide section 2 details the storage & handling guidelines and contaminant exposure to be avoided. The HDC2021's factory-installed polyimide cover tape protects the device during manufacturing and assembly, but after tape removal the sensing element is again sensitive to contamination.
- Can you share the TI lot/batch #? You can share a picture of the label on the original packaging. - Apologies for not having the pictures, but I recall that the batch we received was packaged in a polythene bag containing a black container strip. Sensors were placed inside, each topped with pink foam. We ensured compliance with the datasheet's guidelines for temperature, layout, and soldering. After soldering, we removed the pink foam for testing
- The temperature data fluctuates unpredictably within the chamber and does not align with the Class A sensor readings specified in the datasheet, failing to maintain the required accuracy
- The same observations are observed for the Humidity as Well.