LDC1000 thickness of conductive target

Hello,


I am designing an application in which a very thin target needs to be detected. Is it possible to detect something that is thinner than the skin depth with a high resolution if the metal is very close to the coil (a few mm)?

The skin depth:

for my parameters:

w=2*pi*f=2*pi*5000000

µr=1.000022

rho=2.6548

I get a skin depth of about 36µm. Can my conductive target still be thinner than that?

  • Hello,

    What resolution are you trying to achieve?

    Resolution will also depend on the diameter of the coil.

    Rough estimate: calculate 1-e^(-d/σ) - this is how much your sensitivity will be reduced compared to a thick metal target.

  • Sorry, should have been δ in the last formula. d - thickness of your metal.

  • thickness skin depth sensitivty loss
    10 30 0,283468689
    15 30 0,39346934
    20 30 0,486582881
    25 30 0,565401791
    30 30 0,632120559
    35 30 0,688596776
    40 30 0,736402862
    45 30 0,77686984
    50 30 0,811124397
    55 30 0,840120254
    60 30 0,864664717
    65 30 0,885441156
    70 30 0,903028032
    75 30 0,917915001
    80 30 0,930516549
    85 30 0,941183528
    90 30 0,950212932

    Alright so about 3 times the skin depth for 95% sensitivity. At half the skin depth I lose about 60% of sensitivity.

    I'm trying to measure vibrations in a very thin stretched membrane (a few tens of microns). Displacement of the membrane will only be a few microns, depending on the thickness, stifness and size of the membrane. So ideally my resolution should be as fine as possible (sub micron).

    The membrane will be very close to the coil, probably less than a mm.

    I have another question regarding my application can I ask this in the same thread or do you prefer a new thread?

    Thanks for the help