Elliptic coil

Other Parts Discussed in Thread: LDC1000EVM, LDC1612, LDC1614, MSP430G2553

Hi

I need to sense a stainless vibrating metal piece with height of 6 mm and width of 2 mm. The distance varies between 6-10 mm and I essentially need to see the vibration frequency. This metal piece is immersed in water and the electronics is behind a plastic sheet.

Because the surface area is pretty small the signal is not too strong. I can see the vibrations with LDC1000EVM test coil but I would like to strengthen the signal. Because the metal piece is attached by two very narrow metal legs (length of 20 mm and width of 0.2 mm) If I would make the coil ellipsoidal, say one diameter of 9 mm and the other diameter 15 mm, I probably could couple to signal to the metal legs as well. Are there any considerations that I should take into account if I use ellipsoidal coils?

  • Hello Sami,

    First please make sure that the plastic material between the target and the coils is not a dialectric (some plastics are). Second, I assume you are using L measurement only. In the case, LDC1612 would have a larger response than LDC1000.

    Best Regards,
    Natallia Holubeva
  • I'm currently using proximity data values and compensating with external temperature sensor. The change with the LDC1000EVM PCB coil is around 100 units while the noise level is around 20 units. I cannot see practically any change in the frequency data when using LDC1000EVM GUI. Would the increased resolution in LDC1612 allow me to see changes in the L values? Currently the frequency data varies between 765-767 when the metal plate vibrates which is too small for measurement.

  • Please check the following:

    1) are you using a response time setting of 6144? This is the highest setting and will provide the lowest measurement noise.

    2) The LDC1000EVM has a CLKIN frequency of 6MHz, which is less than the 8MHz maximum frequency. Using 8MHz for the CLKIN frequency will increase the resolution by 33%. Additional techniques to improve the frequency measurement resolution on the LDC1000 is to use a lower frequency sensor and also to ensure you are using the response time setting of 6144.

    As for the LDC1612, yes, you will have significantly improved measurement resolution - more than 8x improvement in frequency measurement (the L measurement - note that the LDC1612 and LDC1614 only measure L and do not measure RP). The LDC1612 can also support a much longer measurement interval, which should also provide more a better meaurement resolution.

    ChrisO

  • Thanks Chris, I will increase the frequency to 8 MHz in our test setup. I will also test if LDC1612 has enough resolution so that we could go to L measurement instead of Rp.
  • Sami,

    I was wondering what is the reson to use proximity data in your application. L measurement is more stable versus temperature so maybe you could switch to L only. What is the oscillation frequency of the sensor? Thank you very much.

    Best Regards,
    Natallia Holubeva
  • I am using proximity data because L data does not show any change when target vibrates, neither in the EVM GUI or in our custom PCB. I am using the coil from the LDC1000EVM and the LDC1000 parameters are below:

    MSP430G2553 clock frequency 12Mhz and LDC1000 CLK 6 MHz
    #define VIV_RPMIN 0x38
    #define VIV_RPMAX 0x17
    #define VIV_SENSORFREQ 0x64
    #define VIV_LDC 0x15
    Target size = 5mm x 2 mm. Distance to target about 8mm.

    Our target is vibrating with varying frequency range between 20 Hz to over 200Hz. I see the vibrations better with LDC of 0x15 when the frequency is higher.
  • Sami,

    Thank you very much for your explanation. Your work is important to us.

    Best Regards,
    Natallia Holubeva