Hi,
My customer needs 0.5s@5m/s response time.
I dont see these units on the datasheet. Is it possible I'm missing this?
Thanks,
EJ
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hi,
My customer needs 0.5s@5m/s response time.
I dont see these units on the datasheet. Is it possible I'm missing this?
Thanks,
EJ
Hi EJ
Not sure to have well understood your question.
No problem to have a 0.5sec period for the conversion. To do this a new conversion has to be triggered every 0.5sec.
Regarding the 5ms response time the situation is the following:
To have an accurate measurement we suggest to convert always both humidity and temperature
The conversion time depends on the selected resolution:
Selecting Humidity 8 bit resolution and Temperature 11bit the conversion time is: 2.5+3.7= 6.2ms
Please let me know if it answer to the customer questions.
Regards
Massimo
We don't have the right set up to test the response time as you request.
Our set up has a fixed air flow at 1m/s.
Best Regards
Massimo
Is there a way to adjust the test to get the information the customer needs?
Thanks,
EJ
Hi
Customer have configured the HDC1000 in both temp & Humidity conversion mode[Writing 0x1000 in to configuration register(address:0x02)], and their feedback as below
1)I am getting 0xFFFC as same value for both temp & Humidity when I am following the below sequence
start,address+w,00(pointer register), stop command, wait for DRDYn to be low , start, address+Read.
2)I am not getting the signal DRDYn asserted if stop command is not given after start,address+w,00(pointer register).
Kindly suggest whether stop command need to be given after the Trigger of start temperature & humidity conversion[by attempting writing 0X00 in Pointer register] before giving the read command of both Temperature & Humidity at a time.
Pls adv
Hi Massimo
Customer have configured the HDC1000 in both temp & Humidity conversion mode[Writing 0x1000 in to configuration register(address:0x02)], and their feedback as below
1)I am getting 0xFFFC as same value for both temp & Humidity when I am following the below sequence
start,address+w,00(pointer register), stop command, wait for DRDYn to be low , start, address+Read.
2)I am not getting the signal DRDYn asserted if stop command is not given after start,address+w,00(pointer register).
Kindly suggest whether stop command need to be given after the Trigger of start temperature & humidity conversion[by attempting writing 0X00 in Pointer register] before giving the read command of both Temperature & Humidity at a time.
Pls adv