Hi TI:
I encountered the following problems in the calibration process
Look forward to your reply!
Best,
Kevin_Wang
Hi Kevin,
This depends on the current settings that you choose and the optics used. Likely multiple targets will be needed because you usually choose the super HDR currents to allow the device to function at a large range of distances/reflectivity.
There are a couple options
1) If you are willing to compromise on accuracy a bit you can decrease the amplitude requirement of 10k slightly without much impact on performance. On the EVM we calibrate phase offset at 2 distances total for the 4 currents. This means the larger of the two currents will be 24k and the lower one will read ~7k. Thus only 2 distances are needed instead of 4.
2) If you are able to vary the target reflectivity then distance need not change.
3) Or you can choose different distances such that amplitude is always 10k or greater for best results.
Best,
Alex
Hi Alex:
I want to know what the reference values of the Iphase and Qphase of xtalk correction are, because our EVM is broken!
Thank you!
Best!
Kevin