TMP112: About TMP112x accuracy

Part Number: TMP112

Hello Expert.

Thank you for your special support every day.

Please tell me about TMP 112x.
There are 3 types of TMP 112A, B, N.
The TMP112x divides the model name according to the accuracy test result.

I have some questions.

- Which of the 3 products is equivalent to TMP112?

- What is the maximum temperature accuracy when TMP 112 A is used at 1.8 V?
  Temperature accuracy maximum value = ± 0.5 ° C @ 3.3 V, 0 to 65 ° C. ± (3.3 V to 1.8 V) * (± 0.25 ° C.) = ± 0.875 ° C.
  Is this a wrong recognition that this is a guaranteed value?

Best Regars.

Noriaki

  • There are additional questions.

    Can I think that all slave addresses of TMP112A, TMP112B, TMP112N are the same specification?

    Best Regars.

    Noriaki

  • Hello Noriaki-san,

    You are correct, the TMP112x family is named according to accuracy, with TMP112A being designed for 3.3V, and TMP112B being designed for 1.8V. 

    To answer your questions:

    - Which of the 3 products is equivalent to TMP112?

    All three are part of the TMP112 family, we could refer to any of them as a TMP112.

    - What is the maximum temperature accuracy when TMP 112 A is used at 1.8 V?

    Your accuracy calculation is correct, 0.25 °C/V is a guaranteed value. So you would not expect to see greater than 0.875 °C inaccuracy if using the TMP112A at 1.8V. Of course we would recommend using the TMP112B if operation at 1.8V is needed.

    - Can I think that all slave addresses of TMP112A, TMP112B, TMP112N are the same specification?

    Yes, all three devices will have the same address selection options.