FDC1004EVM: Huge offset at the Reference Environment (RE) electrode

Part Number: FDC1004EVM
Other Parts Discussed in Thread: FDC1004, TDC1000

Hello,

Im experimenting with the EMV to create a level sensor as described at tidu736a.

We have made a sensor FPC as described at tidu736a, whith the 2 reference electrodes exactly same size.

We set the measurements to differential as described at tidu736a (see screenshot below).

However, I see a very large capacitance of about 4.5 pF at the environment reference electrode (see screensot below). Even when I disconnect the sensor FPC the capacitance is about 2.5 pF. When I switch over to singel ended mode by setting CHB to DISABLED the capacitance is less than 1 pF.

When I put my finger on the environment sensor the capacity further increases, as expected.

I measured the shield and sensor signals with an oscilloscope. I saw that the exitation of the SHLD1 and SHLD2 is in antiphase, as expected. The exitation of the 3 sensors is in phase with SHLD1, as expected.

I noticed that at the FDC1004 datasheet figure 16 also the reference capacity is large, see screenshot below.

Do you have an explanation for this high lreading, is it normal?

How must I take this high value into the fluid level calculation (equation 2 in tidu736a)?

Regards, Paul Borger.

  • It looks as if you are getting similar results as the TI design, and that number should still work with equation 2. Note that you may need to do a total gain adjustment with that equation as well.

    Possibly the additional capacitance is impacted by coupling of the nearby level sensor.
  • Hello Clancy,

    My results indeed look like the figure 16 of the FDC1004 data sheet.

    And of cource, in the formulla any number will give a numeric result, but a large value as measured will result in a negative height.

    Of course, I can play with multiplication factors, which can be negative, to correct this, but I wonder if the desired ambiant compensation still works correctly.

    What I do not understand at all is why this very high value is measured. Any near trace or electrode can of course influence the result, but 4pF? Even the level sensor itselve has less than 1 pF when no fluid is near.

    Remember that when I disconnect my sensor FPC, the value still is about 2pF.

    Attached below I send you pictures of our sensor FPC. It was designed as a very close copy of yours. I wrote the signals on the surface after verifying the connections with a multimeter.

    Meanwhile I noticed also that at figure 3 of tidu736a this high level for the environmental sensor is not seen. Actually this picture is exactly what would be expected. I wonder what you did to achieve this, there is no explanation provided in tidu736a.

    Regards, Paul Borger.

  • P,

    I was able to find one of these sensors, so I am planning on running a similar experiment to verify your results. I will give you another update by the end of the week.
  • P, My apologies for the delay! I am still working on this.
  • Hi Clancy,
     
    OK, understood, no problem.
    Meanwhile I got some progress.
    I think the problem lies in the layout of the small PCB that holds the CDC and the sensor PCB. The channel 3 track runs closely to other signals that influence the capacitance. I could modify the small PCB and noticed significant improvement, but still too much stray capacitance to perform a good measurement. My target has very low Epsilon-r and not conducting, so my full scale readout is only a few pF.
    Now I’m experimenting with ultrasound. TDC1000_7200. First impression is that it works fine, though I noticed that fluid surface must be parallel to the sensor otherwise the reflected wave is not send back to the sensor.
     
    With kind regards,
     
    Paul Borger
    Electronic Engineer | ED-12
     
    +31 6 25655258
    +31 77 359 3337
     
     
     Océ-Technologies B.V.
    P.O. Box 101, 5900 MA Venlo
    Sint Urbanusweg 43, 5914 CA Venlo
    The Netherlands  |  www.oce.com
    Trade register: 12002662
     
    This message and attachment(s) are intended solely for use by the addressee and may contain information that is privileged, confidential or otherwise exempt from disclosure under applicable law. If you are not the intended recipient or agent thereof responsible for delivering this message to the intended recipient, you are hereby notified that any dissemination, distribution or copying of this communication is strictly prohibited. If you have received this communication in error, please notify the sender immediately by telephone or with a 'reply' message and delete the message and its attachment(s), if any, from your system(s). Thank you for your co-operation.
     


  • Yes, the TDC family can be used for liquid level sensing as well, and the signal strength will be strongest when the fluid surface is perpendicular to the ultrasonic sensor.

    If you want more help with that solution, please start a new thread so that our TDC focused engineers can help.