Hi e2e Support:
Please refer to the below figures, we set the hdc2010 sensor board out of case to avoid the inner. All of them has a shift about +5% compared with meter. The meter is calibrated yearly. Additional, we test the hdc2010 EVK. It also has the same offset.
Our product had been using in Japan where the humidity may over 80%RH. And we found that from the spec, "Long exposure outside the recommended operating conditions may temporarily offset the RH output. The recommended humidity operating range is 20 to 80% RH (non-condensing) over 0 to 60°C." And refer to other thread, we got that baking can help to it.
After 36hrs baking at 100℃, these hdc2010 sensor all recovers, but not completely. The shift decrease from +5% to 2.5%.
1. How to deal with the remaining 2.5% shift? Is a calibration OK for it?
2. Do you have any suggestions if the operation environment can exceed 80%RH?
3. We want enable the integrated heater to bake the sensor. But when should it be used? Our plan is when the sensor detect the envirment humidity exceeds 80% for a period of time(T1), then enable the heater for a period of time(T2). Do you have some suggestion about what the T1 and T2 should be?