Other Parts Discussed in Thread: ADS1261, ADS7042
Dear All,
Regarding the output of the pacemaker from the two pins of the AFE ADS1298, TESTP_PACE_OUT1 and TEST_PACE_OUT2.
We are currently working on a detection circuit of pacemaker based on the paper written by Brian Pisani with name TI Precision Designs: Verified Design Software Pacemaker Detection Reference Design.
But instead of using the SAR ADC ADS7042, we are using the ADS1261 with its software to acquire the data to make an algorithm to detect the pacemaker signal from it.
The ECG + PACE signal is obtained from a Whaleteq simulator of model SECG 4.0 to make sure that the detection of pacemaker pulses is done based on IEC standards IEC 60601-2-25, IEC 60601-2-27, IEC 60601-2-47, and its specifications are available on this page: https://www.whaleteq.com/en/Products/Detail/18/SECG%204.0#tag=proOverview.
Here is the schematics for the connection of the input stages of the ECG leads to the AFE, taking in your consideration that the PGA of the AFE is set to 6.
According to the datasheet, to acquire the pacemaker pulses, only one channel must be selected at a time, so we select channel 2 to acquire the pacemaker pulse from the TESTP_PACE_OUT.
The problem appears with the pulses with small values or duration, and here is the spectrum for the samples after performing the filters (Analog and digital) on it.
2mv - 2ms
2mv - 500us
2mV - 100us
As you can see here, the pulse of the pacemaker signal is so small, the 2ms and 500us duration can be detected, but the smallest one (with 100us duration) can not be detected due to its small duration and amplitude.
Also, the signals taken from the other pin (TESTN_PACE_OUT) contains higher noise than the even ones.
Is there any way to reduce this noise, or to detect these small pulses.
Also another question, according to IEC 60601-2-25/27/47, it's required to detect the pace on every single channel, including the chest leads also, in this case how we can configure the PACE Register?, as you can see, the chest leads are connected to the WCT on there negative differential side.
Looking forward to your reply.