OPA2188: Which one is the true value when test the input bias current?

Part Number: OPA2188


Tool/software:

Hi, TI's Masters:

I'm trying to test input current of OPA2188, the test topology is shown below:

I use a  portal multi-meter like FLUKE 289C to measure the voltages of A and B point, then divide the value by 10K, then get the input current of OPA2188.

but here comes a strange phenomenon:

case 1: the black probe contact BAT-, and red probe to A and B respectively, then change the input voltage from 100V to 800V, 100V step, the measure data is as followed below:

input voltage/V voltage A/V voltage B/V
100 0.2005 0.2005
200 0.3871 0.3868
300 0.5737 0.5732
400 0.7604 0.7597
500 0.9476 0.9468
600 1.1342 1.1331
700 1.3215 1.3203
800 1.5090 1.5075

I don't understand why the voltage difference between A and  B become more and more with input voltage rising.

and the V(A-B)=1.5mV at 800V input, so the current across 10K is nearly 150nA, which is ten times more than datasheet.

case 2: I tried to measure A and B immediately, black probe to A and red probe to B, this time the V(A-B) is always nearly 10uV regardless of Vin, less than case 1 much a lot.

So which case is the correct measurement method? and why?

thanks a lot!

  • The Fluke 289 has an input impedance of 10 MΩ; the OPA2188 is specified as 6 TΩ. So most of the current that you're measuring is not the bias current.

  • Hi Xiang, 

    So which case is the correct measurement method? and why?

    Clemens has a good point. You would need to have electrometer or picoammeter to measure the small bias current in the OPA2188. 

    The accuracy of the DMM is not adequate. 10uV is approx. the accuracy limit of the DMM. 

    If you have other questions, please let us know. 

    Best,

    Raymond