ADS1292R: Resistance is not accurate and non-linear across various values

Part Number: ADS1292R

Measuring resistance using ADS1292R. Application is AED device. Using a variable resistor to simulate the PADS resistance when connected to human body. 

Software Impedance = (Average ADC Counts - 525500)/3070

The pads are connected through IN2P and IN2N.

Decade Box (ohms) Average ADC Counts DMM (ohms) Software (ohms)
10 534492 10.2 2
20 565304 20.6 12
30 596400 30.6 23
40 626930 40.8 33
50 657710 50.5 43
60 688572 60.1 53
70 719186 70.6 63
80 749833 80.3 73
90 780217 90.6 82
100 810349 100.1 92
110 841161 110.4 102
120 871205 120.2 112
130 882326 129.8 116
140 909453 140.2 125
150 936528 149.8 133
160 963449 159.4 142
170 990381 170.2 151
180 1016690 179.4 159
190 1042996 189.8 168
200 1070591 200.1 177
210 1096322 210 185
220 1121881 220.1 194
230 1147459 229.9 202
240 1172401 240 210
250 1197390 250.1 218
300 1319643 300.1 258
350 1436212 349.7 296
400 1548883 399.9 333
450 1656890 448 368
500 1760219 498 402
550 1860201 547 434
600 1957423 598 466
650 2050351 647 496
700 2139468 698 525
750 2225924 747 553
800 2309680 797 581
850 2390416 847 607
880 2443559 880 624
890 2458888 890 629
900 2469292 897 633
950 2544756 947 657
1000 2613666 993 680
1050 2684663 1044 703
1100 2762476 1100 728
1150 2828826 1150 750
1200 2892593 1200 771
1250 2955100 1250 791
1300 3014740 1300 810
1350 3078340 1350 831
1400 3134460 1400 849
1450 3190124 1450 867
1500 3243871 1500 885
1550 3296406 1550 902
1600 3346765 1600 918
1650 3396728 1650 935
1700 3446938 1700 951
1750 3493156 1750 966
1800 3539526 1800 981
1850 3584458 1850 996
1900 3628168 1900 1010
1950 3670932 1950 1024
2000 3712686 2000 1038
2050 3752483 2050 1051
2100 3792619 2100 1064
2150 3833009 2150 1077
2200 3870168 2200 1089
2250 3907515 2250 1101
2300 3943747 2300 1113
2350 3978335 2350 1124
2400 4012755 2400 1135
2450 4046337 2450 1146
2500 4080440 2500 1157
2550 4112773 2550 1168
2600 4144331 2600 1178
2650 4175332 2650 1188
2700 4206293 2700 1198
2750 4236075 2750 1208
2800 4265185 2800 1218
2850 4293129 2850 1227
2900 4321436 2900 1236
2950 4348383 2950 1245
3000 4375290 3000 1254
3050 4401432 3049 1262
3100 4429666 3099 1271
3150 4455106 3150 1280
3200 4479771 3200 1288
3250 4504703 3250 1296
3300 4528831 3301 1304

CONFIG REgisters

Final Configuration prior to receiving impedance and ECG data CONFIG1 0x1 0x01 Continuous Converstion Mode, 250 SPS
  CONFIG2 0x2 0xE0 Lead-off comparators enabled, Reference buffer enabled, 2.42V reference, OSC out disabled, Test signal off, Test signal DC
  LOFF 0x3 0x10 Lead-off 15% - 85%, Lead-off current 6 nA, Lead-off frequency DC
  CH1SET 0x4 0x00 Channel enabled, PGA gain is 6, Normal electrode input
  CH2SET 0x5 0x40 Channel enabled, PGA gain is 4, Normal electrode input
  RLD_SENS 0x6 0x20 Chop frequency Fmod/16, RLD buffer enabled, RLD lead-off sense disabled, CH2 RLD negative inputs not connected, CH2 RLD positive inputs not connected, CH1 RLD negative inputs not connected, CH1 RLD positive inputs not connected
  LOFF_SENS 0x7 0x00 FLIP2 disabled, FLIP1 disabled, LOFF2N Disabled, LOFF2P Disabled, LOFF1N disabled, LOFF1P disabled
  LOFF_STAT 0x8 0x40 Fmod divider is 16, RLD is connected, IN2N_OFF connected, IN2P_OFF connected, IN1N_OFF connected, IN1P_OFF connected
  RESP1 0x9 0xEA RESP_DEMOD_EN1 ON, RESP_MOD_EN ON, RESP_PH is 112.5 degrees, RESP_CTRL internal clock
  RESP2 0xa 0x01 Calibration off, RESP_FREQ is 32kHz, RLDREF external