I was testing my circuit with sigma delta and comparing with an SAR architecture design. I am running both at a 20000 sps. So I was expecting a group delay of 38/20000 = 1.9mS for all the frequenciesin sigma delta. But as I started testing the variation in delay with the frequencies. I am getting strange results. At lower frequencies(10-100hz), I am getting the theoretical delay of 1.9mS and at higher frequencies (>500hz) I am not getting any delay. I still dont have the data between 100hz and 500hz. I will let you know as my testing progress. The group delay due to analog amplifiers and filters on both SAR and sigma delta is less than 100 micro second.
Test setup: waveform generator : 33500B series configured to differential signal output using two channels combined. Both devices connected in parallel.
In pulse testing, both sigma delta have lost 50mS of data, which is also unexpected.
What am I missing here? Is this phenomenon anyway related with sigma delta group delay/ settling time issue ?
DATA: pdf
I found some of the legends are not proper. Please refer below:
For all graphs Xaxis is time (mS)
Y axis: Counts
Blue color graphs are: SAR
Orange graphs are sigma delta.