Hi ,
I am working on wireless heart monitor using ADS1293 with CC2541.Everything went well until I adjusted ODR. When increase the ODR I got waveform distortion.
Here is my test data:
It's OK when ODR=160Hz
Waveform distortion when ODR=320
Can anyone kindly help me to find out the reason for waveform distortion in higher ODR.
Many Thanks,
Ru Hua
From the posted data, apparently CC2541 is not catching up with the selected higher data rate programmed in the ADS1293. You may have to run your system at a lower rate to avoid the distortion in the signal.
Hello,
You should be able to connect a function generator directly to the inputs. You need to make sure that your signal at the pins of the ADS1293 are less than the maximum allowed signal and with the CMVR. This is explained in section 8.3.5 of the datasheet.
Mike