Hello,
We are using ADS1298IPAG in our ECG signal acquisition project. I am trying to improve the CMRR as per suggestions in the application note SBAA188–July 2011 " Improving Common-Mode Rejection Using the Right-Leg Drive Amplifier". I was refering figure.7 in this document to find Cp value. It is mentioned "The value of CP is chosen such that zero as a result of RP and CP is approximately 60 Hz" I have chosen RP value 50K and calculated CP value as 53nF. Is this calculation correct? If not, can you share the right formula?
thanks,Shanoj
Hi Shanoj,
Thanks for your post!
This post should help: https://e2e.ti.com/support/data-converters/f/73/t/747501