Hi TI team:
Please help to confirm the attachment problem
This thread has been locked.
If you have a related question, please click the "Ask a related question" button in the top right corner. The newly created question will be automatically linked to this question.
Hi Chunqiu,
Please see this post: https://e2e.ti.com/support/data-converters/f/73/t/799662
I would recommend trying a configuration more similar to this one. Currently, you do not have the AC blocking capacitor in the line.
Hi Chunqiu,
In addition to Alex's comment, I would not use both methods (1.) and (2.) simultaneously. Method (2.) with the 10M pull-up to RLD will help a little more to maintain CMR, but IN2P and IN2N must be enabled in the RLD_SENSn registers.
Best regards,