Hi. I am designing a system for reading a large number (~600) of thermocouples using multiple modules. Each module using the ADS124S08 reads out 5 thermocouples and one thermistor (for CJC). Each of the thermocouples is grounded, and all 5 of the thermocouples transition from thermocouple wire to copper wire at an insulated isothermal junction block, so all 10 thermocouple connections are at the same temperature, and that temperature is read by the one thermistor. The raw voltages are sent upstream to a larger system which calibrates the data and converts it to temperatures.
I've designed a system using an isolated power supply derived from an ISOW7841 isolator, and an LM27762 regulator/charge pump to prevent a short circuit from 277V power in the furnace to one of the thermocouples from destroying more than one of the modules, in the worst case. An LM 27762 charge pump/regulator is used to derive a bipolar supply so that the grounded thernocouple voltages fall in the middle of the power supplies.
I have a few questions about the design, especially about the grounding. I'll attach a simplified schematic of the ADC section with one of the thermocouple circuits shown.
1) Should I solidly ground the middle point of the +-2.5V supplies, and the digital ground? Or should I use a high-impedance AC/DC ground between the signals and the earth ground, bearing in mind that the earth ground is tied solidly to my thermocouples.
2) I am thinking I should put clamp diodes on the signals after they go through the fuse and the 2770 ohm resistor to clamp any short circuit currents to the rails to protect the ADC inputs.
3) Is there any point in using an isolated power system if the signals are grounded anyway? I was thinking that it would reduce the potential for ground loops, but I don't know if that's true or not.
Any other observations about the design would be welcome!