EVM430-FR6047: Characterization

Part Number: EVM430-FR6047

I finally resolved how I should perform the VFR calibration. 
Now we have to get about 5 more tubes with sensors to have statistical data about their operation.
I have some doubts that I would like to clarify.
We are using tubes from the audiowell company, particularly the model hs0014-000. 
I wanted to know if you have any characterization curve (vfr + temperature) for these tubes.
My question is when I have characterized a particular tube, is that characterization the same in all the other tubes of that same brand and model?. 
In the production process should we load that characterization curve and perform a calibration of the tube constant?
  • Hi Jucetauro,

    We do not offer a characterization curve for any devices.

    Once you have characterized a particular tube, the characterization from one compared to another of the same model should be quite similar. In production, we would expect that you can load this curve to all devices of a particular model.