Part Number: EVM430-FR6043
Tool/software:
We have developed a sensor that is used to measure flow for fluids between, roughly, 0.1 and 10.0 lpm with a circular cross-section of about 71 mm² at a temperature that varies around room temperature. We have worked out how to calibrate it for a particular scenario, e.g., water at room temperature. The issue is we do not necessarily know, a priori, what fluid will be measured in the field.
The general question is this: is it possible to calibrate at the factory, establishing certain sensor stable characteristics such as ToF under known conditions, and then "correct" in the field using any observed changes between those established characteristics and the currently observed values to report an accurate flow rate for the current fluid? Specifically:
- What should the established parameters be?
- What should the correction approach be?
- Should any parameters be changed in USSConfig or can the correction be done downstream of USS flow rate output?
- How should non-fluid ToF, i.e., unchanging with respect to fluid changes, be considered? Ours is something like 17% of total ToF.
We've considered and tested a number of approaches/formulas, but none of them reliably yield the desired result.
For extra credit, what unit is USS_VOLUME_SCALE_FACTOR in, and how would one establish a sane value using the known physics and geometry of the sensor? While we do establish a value during calibration that yields good measurements, the units seem arbitrary.
Thanks in advance.