MSP430FR6043: MSP430FR6043 error

Part Number: MSP430FR6043


Hi TI team,

Does the USS library takes care of temperature compensation in metrology calculations results

Do we require to separately compensate this error and How?

Secondly upto what temperature it is compensated.

Does the library also measure water temperature?

Thanks

  • Hi Virendra,

    The temperature can be determined from the absolute time of flight(using the known speed of sound in water).  A separate temperature sensor is not required.  Volumetric flows can also be calibrated across temperature ranges to give more accurate results using the calibration tab in the GUI. We typically test our tubes from 5C to 85C.

    BR,
    Leo