What is the appropriate grounding scheme for a Right-leg drive? I was refering to the link below:
http://www.ti.com/lit/an/sbaa188/sbaa188.pdf
This assumes that the person is not connected to earth.
If the system is operated on 0-5V single supply, then if the person inadvertently touches earth, like operating an equipment or simply walking on the floor, then would not the RLD be a problem? Considering a proper electrode connection to body, body impedance can be ~500 ohms. So, if the DC feedback by RLD is 2.5V w.r.t. electrical ground, and I assume the electrical ground ~= earth potential, then 2.5V/500 ohm = 5mA flows through the body, which is close to the macro-shock limit!
Please correct my above understanding.
Or is there any particular note on the grounding scheme? or some other electronics should be added to account for the above safety?
Also, is there any application note from TI regarding the safety limits/procedures for biomedical electronics?