Hello everyone,
I'll just repost a question I asked earlier in an already answered thread:
Using the EVM I found that it shows approx. 50nF with nothing connected and approx. 140 nF with my testing equipment (wiring, oscilloscope etc.) connected, but still without the sensors. This all makes sense and the corresponding frequencies in the range of 3 MHz shown on the oscilloscope are as expected.
However, my sensors are of variable capacitance in the range of .66 to 7.2 nF. This is where the problem starts. If I connect a 33nF film capacitor to the EVM, the frequency shoots up to 10 MHz instead of significantly decreasing. The oscilloscope and the GUI both show the same frequency, so the frequency measurement is correct.
Can someone explain what happens here?
I suspect it has to do with the energy stored in the capacitance (E=1/2 * C * U^2) and in the inductance (E=1/2*L*I^2). Since the oscillation is in effect the energy swinging back and forth between the two, the energy stored at peak must be the same in both L and C.
I can measure the maximum voltage of the oscillation (in the range of 1.2 V to 1.8 V, as per data sheet) and thus know about the energy stored in the capacitor. What about the current? I am unclear of how the current oscillating between L and C is related to the DRIVE_CURRENT set in the GUI. Can someone enlighten me?
Summary: I suspect that if my sensor has a large capacitance, I also need a large inductor. Is there some guidance from TI how to set the value of L for a given range of C, defined by the physics of my sensor?
Thank you in advance for helping me out.