AFE4405EVM: auto calibration

Part Number: AFE4405EVM

Hi, 

I am going to get PPG data after applying auto calibration. It seems that the firmware for MSP430 of AFE4405EVM supports calibration

but the current PC GUI app doesn't have a way to send a command to activate it.

To make the calibration work ASAP, which is the best way ? Please advise on the following options.

1. A next version of PC GUI app is available, which supports it.

2. Just modify & rebuild the MSP430 firmware to support it using CCS. The start of the calibration will be triggered by some other command from the GUI. (No change in the PC GUI app.)

Regards,

Konan Kim

  • Konan,

    I'm assigning your post to the correct applications engineer. You should get a response soon.
  • Hi Amy,

    Thanks for your reply. In the meantime, I've prepared the environment to modify the firmware of the EVM board by using CCS. Furthermore, I've found that the current GUI supports "calibration enable" under the ADC capture & analysis tab.

    The problem is that even though I run the calibration according to the guide which pops up when the "enable calibration" is checked, the result is a failure.

    Here are the steps.

    1. After the "enable calibration" is set, the capture started. The screen shows LED1 is fixed at 0.3V (all the time) and other LEDs show noisy signals. (The 1st picture) From the firmware, I can see that LED1 is default target for calibration.

    2. After the step, read all the registers into the GUI. And then unchecking the "enable calibration", capture started. The result is that LED1 showed a saturation or very noisy signal which are near 1.0V even though the calibration is done. My expectation is that the DC values of the resultant LED1 signal is around 0.4 ~ 0.6 volt or simply 0.5 volt.

    3. The following pictures show that the parameters after the calibration. The problem is that the current value of LED1 is too small(0.782mA) and the gain is large which will make noisy signals. This results was measured on wrist. Tried many times but the results were the same. Even tried at fingers, the results were the same.  To overcome this, if some initial values need to be modified in the firmware, please let me know. I can modify and rebuild the firmware now.

    4. The sampling freq and duty cycles were 100Hz  and 3% respectively.

    Regards,

    Konan

  • Hi Konan,

    Let me check the firmware and get back to you offline.

    Regards,
    Prabin