Tool/software:
Hi TI team,
Recently our production team facing a few issues on DDS LO board at cold temperature chamber (0°C) which is the alignment failure occur on DUT at first boot up at cold. The alignment shows the calibrator level signal is too low.
Symptoms:
- Chamber is seeing spikes in failures with multiple alignment failures starting with AifVariableAttenuatorAlign.

- When failing, both Ecal and the "front end" paths see 50Mhz signal low, @ ~ -55dBm at COLD conditions


- The Cal signal tends to recover with alignment run when the temperature gets higher.
- Signal was measured solid out of Ref Plus, but low out of LODDS output. The fact that Ecal also failed narrowed the scope down to few boards, primarily LODDS.
The failure can be duplicate when install the defect component on another good DDS LO board. And our production tried to swap the component by take from the good old DDS LO board and from the bin stock, and both are pass.
There is a body marking on failure component which is A BB 0P A 2227U K 39APGN

Block Diagram

Defect part was sent to FA, and TI conclude TNI (Trouble Not Identified). They confirm these return units are within the data sheet specification for IQ parameters, but for applications which is failure abnormal signals require application review.

Please kindly help to advise.
Regards,
Amira