Part Number: INA238
Tool/software:
for my aplication: maximum expected current= 90A, ADCRange=0,(163.84mV), Rshunt=0.001 Ohm
This is my messure results up to 10A. It seems that as the current increases, the measurement error becomes larger.
| is | INA238 | |
| 0 | -2,75 | |
| 505 | 524 | |
| 1010 | 1052 | |
| 1515 | 1579 | |
| 2019 | 2106 | |
| 2322 | 2425 | |
| 2524 | 2633 | |
| 3028 | 3164 | |
| 3533 | 3688 | |
| 4030 | 4218 | |
| 4540 | 4746 | |
| 5040 | 5273 | |
| 5550 | 5800 | |
| 6050 | 6328 | |
| 6560 | 6858 | |
| 7065 | 7382 | |
| 7570 | 7915 | |
| 8080 | 8445 | |
| 8590 | 8973 | |
| 9090 | 9500 | |
| 9560 | 10027 | |
| 10000 | 10555 | |
| 10500 | 11085 | |
| 11000 | 11618 | |
| 11500 | 12150 | |
| 12000 | 12680 |
Why is there such a large error? Can I easily calibrate it with a proportional factor?
2. in Datasheet, to caculate current_LSB, the Maximum Expected Current is in my sitiation 90A? Why is this parameter so important? for exsample, when my Expected Current is 100 A, the Current_LSB will change.
or is Maximum Expected Current a calculated parameter, according to Rshunt. or it is only a Estimated parameter?