Part Number: LM6171
Summary from customer:
Error 7 is an error upon self-test. The device's stimulation output is verified before treatment. If it falls outside of a designated range, it returns Error 7. The schematic I attached has the stimulation generation circuit on page 4 and the parameter sensing circuit on page 5.
EWME3525-0035, EWME3525-0002, EWME3525-0016
0035 and 0002 came from the field. The analysis returned the following results:
0035
- Upon visual inspection, capacitor C36 was visibly bulging.
- Both current and voltage were below the threshold on self-test
- Upon installing the PCBA in a test unit and turning it on, U17 rapidly heated up.
- Replacing C36 and U17 fixed the PCBA and made it functional again.
0002
- Upon visual inspection, capacitor C36 was visibly bulging.
- Both current and voltage were below the threshold on self-test
- Replacing C36 and U17 fixed the PCBA and made it functional again.